hello,
I am new to BioSemi and I have some basic questions that I would appreciate if you can help me.
I am trying to use my own program to acquire the data form the biosemi and then send them to actiview.
I would like to know though, which exactly transformations are performed in Actiview in order to pass from the digital data with dig.max=8388607 and dig.min=-8388608, to the values in actiview with phys.max=8388 and phys.min=-8389, and then from those how we pass to the normal values, for example for temperature ( normal for epidermal temperature==around 32 celcius), that are depicted as an output to actiview.
I am working on an affective game and I am recording peripheral signals.
My goal is to aquire the data with biosemi pass it to actiview through tcp/ip connection and then according to the values acquired that indicate stress or boredom adjust the level of the game.
I know that my question is basic, but i need some help to understand!
Thank you.
Konstantina
Biosemi & Actiview values--peripheral signals
-
- Site Admin
- Posts: 1141
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Biosemi & Actiview values--peripheral signals
Via TCP/IP you receive 24-bit samples (range is from -8388608 to -8388607).
Use the following scaling factors:
- Electrode channels: LSB = 31.25 nV (divide the 24-bit number by 32 to get a value in microVolt)
- Temperature channel: LSB = 1 millidegree Celsius (divide the 24-bit number by 1000 to get a value in degree Celsius)
Best regards Coen (BioSemi)
Use the following scaling factors:
- Electrode channels: LSB = 31.25 nV (divide the 24-bit number by 32 to get a value in microVolt)
- Temperature channel: LSB = 1 millidegree Celsius (divide the 24-bit number by 1000 to get a value in degree Celsius)
Best regards Coen (BioSemi)
Re: Biosemi & Actiview values--peripheral signals
Thank you so much for your instant reply.
One more question.
What for the GSR electrode?
I suppose I also have to divide with something and then proceed as described here?
viewtopic.php?f=7&t=400&p=1011&hilit=he ... alue#p1011
Thank you.
Konstantina
One more question.
What for the GSR electrode?
I suppose I also have to divide with something and then proceed as described here?
viewtopic.php?f=7&t=400&p=1011&hilit=he ... alue#p1011
Thank you.
Konstantina
-
- Site Admin
- Posts: 1141
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Biosemi & Actiview values--peripheral signals
Use the following scaling factor for the GSR channel:
LSB = 31.25 picoSiemens (divide the 24-bit number by 32 to get a value in nanoSiemens)
Best regards, Coen (BioSemi)
LSB = 31.25 picoSiemens (divide the 24-bit number by 32 to get a value in nanoSiemens)
Best regards, Coen (BioSemi)
Re: Biosemi & Actiview values--peripheral signals
Dear Coen,
Thank you for your reply.
I just need some further explanations and your help.
As explained before I use the .bdf data to control my system which is a Tetris game.
The control of the system is going to happen online, that means that I have to compute whether the player is stressed or bored according to his physiological values and then adjust the system according to his state.
Therefore I need to be very precise about the values so that I can compare them to the physiological values that exist in the literature and then proceed to the adjustment of the game’s level.
If I understand well, the relation between the digital values and the physical values is a linear system.
So for a given digital value we can compute the physiological value.
The solution of this system, as I computed, gives an equation and for example for the temperature the solution is:
Real value=-0.5+ 1000*digital value.
In your post before, you suggested that for temperature I should divide all the values with 1000 though, in the solution I computed there is a difference of -0.5 which in case of body temperature is an important difference.
Please tell me If I am wrong on that and what should I do for the real value computation.
Thank you in advance,
Konstantina
Thank you for your reply.
I just need some further explanations and your help.
As explained before I use the .bdf data to control my system which is a Tetris game.
The control of the system is going to happen online, that means that I have to compute whether the player is stressed or bored according to his physiological values and then adjust the system according to his state.
Therefore I need to be very precise about the values so that I can compare them to the physiological values that exist in the literature and then proceed to the adjustment of the game’s level.
If I understand well, the relation between the digital values and the physical values is a linear system.
So for a given digital value we can compute the physiological value.
The solution of this system, as I computed, gives an equation and for example for the temperature the solution is:
Real value=-0.5+ 1000*digital value.
In your post before, you suggested that for temperature I should divide all the values with 1000 though, in the solution I computed there is a difference of -0.5 which in case of body temperature is an important difference.
Please tell me If I am wrong on that and what should I do for the real value computation.
Thank you in advance,
Konstantina
-
- Site Admin
- Posts: 1141
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Biosemi & Actiview values--peripheral signals
If you need a very accurate temperature readout, then you should calibrate the system with a reference thermometer.
Best regards, Coen
Best regards, Coen