Hello,
I have a few questions in interpreting aux sensor channels from a dataset with 16 aux channels, any advice is appreciated:
1) Is this still the correct way to process GSR signals https://www.biosemi.nl/forum/viewtopic. ... nsor#p2450
2) After scaling resp and pleth, are the units simply uV (no conversion to, for example, mm displacement in resp belt)
3) I'm having a hard time arriving at reasonable temperature values, is this an accurate translation of what happens in ActiView?
signal1 = double(Non_Inverting_Temp - Inverting_Temp) + 2*2^30
a = -1302781; b = 157168911; c = -3342225101;
signal2 = sqrt(4*a*signal1 + b^2 - 4*a*c)
signal3 = (signal2-b)/(2*a)
signal4 = int(signal3*1000)*256
signal5 = double(signal4)/8192*32/1000 (conversion from uV to degC)
Thanks,
Clement
Processing sensor channels
-
- Site Admin
- Posts: 1126
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Processing sensor channels
1) Yes, GSR processing has not changed
2) Yes, RESP and PLETH are in uV, but not calibrated.
3) First step is a little different:
Signal1 = double(Non_inverting_Temp+2^30) - (Inverting_Temp-2^30)
If I remember correctly (it's 20 years ago) the signal (Non_Inverting_Temp - Inverting_Temp) could overflow integer32 values and therefore the adding/subtracting of 2^30 was added before the differential-to-single-ended step.
Best regards, Coen (BioSemi)
2) Yes, RESP and PLETH are in uV, but not calibrated.
3) First step is a little different:
Signal1 = double(Non_inverting_Temp+2^30) - (Inverting_Temp-2^30)
If I remember correctly (it's 20 years ago) the signal (Non_Inverting_Temp - Inverting_Temp) could overflow integer32 values and therefore the adding/subtracting of 2^30 was added before the differential-to-single-ended step.
Best regards, Coen (BioSemi)
Re: Processing sensor channels
Thanks Coen! I found the data I was working with scaled the aux sensor channels the same way as EEG channels, so I had to first recover the 32bit words. Everything works now.