Hi,
I am currently helping to put together an EEG task for infants of 12 months. I only have experience with biosemi and adults and I was wondering what were the recommended configuration settings in actiview for infants of that age - the sampling rate, filters, etc. Also what is recommended for referencing? Any advice here would be appreciated.
Thank you.
infant research
-
- Site Admin
- Posts: 1152
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: infant research
Everything is the same as with adults, just purchase the appropriate small sized headcaps, see https://www.biosemi.com/headcap.htm
Since the electrodes are closer to together, there is more change on gel bridges. So, be careful to not use too much electrode gel (just enough to fill the holder).
Best regards, Coen (BioSemi)