Synchronizing EEG signal with EMG and speech

Post Reply
mfn
Posts: 2
Joined: Wed Jun 23, 2021 3:38 pm

Synchronizing EEG signal with EMG and speech

Post by mfn »

Hello,

We were wondering if you could help us with some ideas.

For en EEG project, we would like to integrate and synchronize the Biosemi EEG recording with other signals, specifically with EMG -from the wireless Delsys system- and speech, recorded with a microphone.
Do you know how is it possible to do this?

Thank you so much for your help,
Mihaela

Coen
Site Admin
Posts: 1014
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Synchronizing EEG signal with EMG and speech

Post by Coen »

The ActiveTwo system can measure EEG, EMG (with flat electrodes and headbox, see https://www.biosemi.com/accessoires.htm) and microphone (via ERGO input), all simultaneous, synchronized and up to 16 kHz sample rate. Trying to use other devices for EMG and sound acquisition seems a needless complication.

Best regards, Coen (BioSemi)

mfn
Posts: 2
Joined: Wed Jun 23, 2021 3:38 pm

Re: Synchronizing EEG signal with EMG and speech

Post by mfn »

Thank you very much for the answer!
Best regards,
Mihaela

Post Reply