- Mon Apr 01, 2019 3:23 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Matlab TCP Connection
- Replies: 4
- Views: 15057
Re: Matlab TCP Connection
1) Both computers should be connected to the same network. The transmission speed does affect lag. 2) Select the "Sent local Port" in ActiView. Ounce the connection is established, ActiView displays the "Remote Display Port" with the same number. 3) Check the select box "Add 7 Sensors". RESP is the ...
- Fri Mar 22, 2019 1:08 pm
- Forum: BioSemi Hardware
- Topic: Fault with ActiveTwo AD Box
- Replies: 4
- Views: 11870
Re: Fault with ActiveTwo AD Box
A gel bridge in the AD-box connector is the most likely cause.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Mar 22, 2019 12:26 pm
- Forum: BioSemi Hardware
- Topic: Pin Electrode Tip Broke Off
- Replies: 1
- Views: 8976
Re: Pin Electrode Tip Broke Off
Pins break due to wear, aging and rough handling. Usually, performance is hardly affected (check noise in water). In case of a single broken pin, we can replace the active electrode (sent set to BioSemi). Replacing electrodes is generally not worthwhile on well used sets.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Mar 22, 2019 11:52 am
- Forum: BioSemi Hardware
- Topic: Fault with ActiveTwo AD Box
- Replies: 4
- Views: 11870
Re: Fault with ActiveTwo AD Box
Dried gel is often invisible. Your top connector board needs to be replaced. Contact BioSemi for service and sent your AD-box.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Mon Mar 04, 2019 1:26 pm
- Forum: BioSemi Hardware
- Topic: Odd ~17Hz noise in all channels
- Replies: 9
- Views: 17121
Re: Odd ~17Hz noise in all channels
Interference by GSR is usually Common Mode, and can be reduced substantially by referencing (which should always applied anyway before analysis). However, in a FFT over a lot of points, the 16 Hz peak will of course eventual always show up, this is normal. Ensure a good (equal) electrode contact of ...
- Tue Feb 26, 2019 4:07 pm
- Forum: BioSemi Hardware
- Topic: Problem with 32 channel headbox / EKG
- Replies: 6
- Views: 15301
Re: Problem with 32 channel headbox / EKG
>Question 1 : >I think Amp / ADC modules are not installed on the C channels. >Could you please explain me how to do it (level : beginner) ? See my earlier post in this thread about hwo to check which channels are installed. Contact BioSemi directly for purchase and installation of more channels. >Q...
- Mon Feb 25, 2019 12:53 pm
- Forum: BioSemi all other topics
- Topic: ECG using Flat Active electrodes
- Replies: 1
- Views: 10786
Re: ECG using Flat Active electrodes
An option is to measure ECG between the arms, search the web for "Einthoven lead I".
Best regrads, Coen (BioSemi)
Best regrads, Coen (BioSemi)
- Sun Feb 24, 2019 10:59 pm
- Forum: BioSemi all other topics
- Topic: Differences between decimater utility and Python
- Replies: 3
- Views: 13220
Re: Differences between decimater utility and Python
The decimator tool does just apply a 5th order cascaded integrator-comb (CIC) filter, no extra filtering/smoothing. Inspect the (open-source) LabVIEW code for confirmation.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Thu Feb 14, 2019 1:10 pm
- Forum: BioSemi all other topics
- Topic: Missing triggers and data
- Replies: 6
- Views: 16603
Re: Missing triggers and data
> 1. Why were these channels recorded in the first place? What did I need to do for them not be recorded? Use the proper setting in the ActiView file dialog. I assume you use 4 EX channels for EMG. 1) Select to only display the 4 EMG channels in the monopolar display (use Channel selection option "f...
- Mon Feb 11, 2019 4:22 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: ActiView cfg file format documentation
- Replies: 1
- Views: 10995
Re: ActiView cfg file format documentation
The CFG is meant to be edited from ActiView. Just a few entries should be edited directly (e.g. with Notepad or Wordpad). For more info, see viewtopic.php?f=2&t=7&p=2232
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Tue Feb 05, 2019 3:15 pm
- Forum: BioSemi all other topics
- Topic: Missing triggers and data
- Replies: 6
- Views: 16603
Re: Missing triggers and data
All I know about Analyzer is that it was updated with a total BDF blockade as far back as 2003 (after working perfectly between 2000 and 2003). Since then, all relevant publications from the vendor list the statement "BioSemi not supported". Not surprisingly, most of the BioSemi customers were quite...
- Mon Feb 04, 2019 3:21 pm
- Forum: BioSemi all other topics
- Topic: Missing triggers and data
- Replies: 6
- Views: 16603
Re: Missing triggers and data
Do you see the same data corruption in ActiView during recording ?
Do you see the same result in alternative BDF file readers such as https://www.biosemi.com/download/ToolsF ... er-808.zip and http://www.teuniz.net/edfbrowser/ ?
Best regards, Coen (BioSemi)
Do you see the same result in alternative BDF file readers such as https://www.biosemi.com/download/ToolsF ... er-808.zip and http://www.teuniz.net/edfbrowser/ ?
Best regards, Coen (BioSemi)
- Mon Feb 04, 2019 3:15 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Matlab TCP Connection
- Replies: 4
- Views: 15057
Re: Matlab TCP Connection
TCP connection is not available on all speedmodes. If the TCP tab is greyed-out during running, then you are using a speedmode where TCP is unavailable (multibox, AIB). Test with speedmode 4,5,6 or 7. Also see viewtopic.php?f=2&t=413&p=3007
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Feb 01, 2019 12:35 pm
- Forum: BioSemi all other topics
- Topic: ecg
- Replies: 6
- Views: 15455
Re: ecg
Order "CMS/DRL Flat Set" (EUR 100 excl. shipping). First check that your AD-box has a CMS/DRL connection on the front-panel (there should be a CMS/DRL label above the most left DIN connector). If not, order "CMS/DRL Flat Set SCSI".
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Thu Jan 31, 2019 2:43 pm
- Forum: BioSemi all other topics
- Topic: ecg
- Replies: 6
- Views: 15455
Re: ecg
CMS/DRL is required for any electrode measurement. When not using the headcap, use a seperate (flat) CMS/DRL connected to the front-panel of the AD-box.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Jan 25, 2019 3:17 pm
- Forum: BioSemi Hardware
- Topic: Trigger inputs through USB receiver
- Replies: 1
- Views: 8583
Re: Trigger inputs through USB receiver
> My question is will the actiview software detect the trigger signal when pin1 goes to 5V.
Yes.
> I believe biosemi recognizes that the trigger pins are 'active high' when there is no input. should i ground the other trigger pins?
Not necessary.
Best regards, Coen (BioSemi)
Yes.
> I believe biosemi recognizes that the trigger pins are 'active high' when there is no input. should i ground the other trigger pins?
Not necessary.
Best regards, Coen (BioSemi)
- Tue Jan 22, 2019 4:07 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Trigger code oddities
- Replies: 6
- Views: 14998
Re: Trigger code oddities
As instructed above, use "analog" trigger display mode to see which bits (besides LSB for 0 -->1) are switching.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Sat Jan 19, 2019 10:26 am
- Forum: BioSemi Software (ActiView executable)
- Topic: Trigger code oddities
- Replies: 6
- Views: 14998
Re: Trigger code oddities
Values are listed in the post link given above. Return to 0 after each trigger.
Bets regards, Coen (BioSemi)
Bets regards, Coen (BioSemi)
- Mon Jan 14, 2019 3:09 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Trigger code oddities
- Replies: 6
- Views: 14998
Re: Trigger code oddities
Instead of 0 --> 255 -->0 , your code probably generates 0 --->255 ---> 254 ---> 0 0 = 00000000 254 = 11111110 255 = 11111111 The 8 bits do not return to 0 simultaneously. Solution: toggle only one bit per trigger event, see http://www.biosemi.nl/forum/viewtopic.php?f=7&t=1022 In general: always use...
- Thu Jan 10, 2019 1:09 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 3
- Views: 11906
Re: Error 5000
Common causes of intermittent 5000 errors are (ordered from most to least likely): - bad optical fiber - bad USB cable - bad optical transmitter - bad optical receiver Replace fiber and USB cable. If the problem persists, then sent the AD-box and USB receiver to BioSemi for testing and repair. Best ...
- Fri Jan 04, 2019 1:04 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 1
- Views: 9555
Re: Error 5000
Use the registry cleaning tool to remove the manually installed driver, see https://www.biosemi.com/faq/install_USB.htm
Do not install the driver manually under W10, just plug in the USB-Receiver and the driver will install automatically.
Best regards, Coen (BioSemi)
Do not install the driver manually under W10, just plug in the USB-Receiver and the driver will install automatically.
Best regards, Coen (BioSemi)
- Tue Dec 18, 2018 2:46 pm
- Forum: BioSemi all other topics
- Topic: Calculating CMS & DRL Values
- Replies: 2
- Views: 12359
Re: Calculating CMS & DRL Values
The data stream consists of measured voltages. A voltage is a difference in potential. In case of the ActiveTwo, the electrode potentials are measured with respect to the potential of a point between the CMS and DRL electrodes, see http://www.biosemi.nl/forum/viewtopic.php?f=1&t=177&p=391 . You are ...
- Wed Dec 12, 2018 1:36 pm
- Forum: BioSemi Hardware
- Topic: Odd ~17Hz noise in all channels
- Replies: 9
- Views: 17121
Re: Odd ~17Hz noise in all channels
Is GSR used and connected to the subject ?
Is the interference seen in the bucket test (without GSR electrodes) ?
Best regards, Coen (BioSemi)
Is the interference seen in the bucket test (without GSR electrodes) ?
Best regards, Coen (BioSemi)
- Wed Nov 28, 2018 12:42 pm
- Forum: BioSemi all other topics
- Topic: Trigger error in low sample rate
- Replies: 1
- Views: 11412
Re: Trigger error in low sample rate
Use positive trigger logic (down-sampling uses OR function to calculate trigger of multiple samples)
Use "analog" trigger display to investigate the problem.
see viewtopic.php?f=7&t=1022
Best regards, Coen (BioSemi)
Use "analog" trigger display to investigate the problem.
see viewtopic.php?f=7&t=1022
Best regards, Coen (BioSemi)
- Tue Nov 27, 2018 5:13 pm
- Forum: BioSemi Hardware
- Topic: Custom electrodes for EXG channel
- Replies: 1
- Views: 8744
Re: Custom electrodes for EXG channel
BioSemi treats the schematics as proprietary information, they are not shared with anyone outside the company.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Mon Nov 05, 2018 3:13 pm
- Forum: EEG data analysis
- Topic: Impact of reference electrode for import on data ?
- Replies: 3
- Views: 13710
Re: Impact of reference electrode for import on data ?
It does not matter. A reference far away generates high Common Mode voltages, a reference close to the area of interest generates small signal amplitudes. Re-referencing cancels these effects.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Nov 02, 2018 12:49 pm
- Forum: EEG data analysis
- Topic: Impact of reference electrode for import on data ?
- Replies: 3
- Views: 13710
Re: Impact of reference electrode for import on data ?
A voltage is a difference in potential. So, the choice of reference does of course affect the eventual EEG voltages. The choice during import does not affect later alternative choices of the reference.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Thu Nov 01, 2018 7:15 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: modify STRIDE value for use with Matlab API
- Replies: 1
- Views: 10205
Re: modify STRIDE value for use with Matlab API
The default Stride is set at approx. 60 milliseconds. If your ("huge") delays are much larger than this value, then the bottle-neck lies elsewhere (ActiView, TCP/IP transfer, PC processing speeds, etc.). Stride can be adjusted by calling functions in the Labview_DLL.dll: OPEN_DRIVER_ASYNC --> BSIF_S...
- Wed Oct 31, 2018 4:24 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Sending Frequency of data from TCP/IP
- Replies: 7
- Views: 15959
Re: Sending Frequency of data from TCP/IP
Your approach will work, providing that the value generated by ActiView divided by the value you select manually is an integer. For example: if ActiView generates a value of 16 samples/chan, the you can manually set 1, 2, 4 or 8 samples/chan.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Oct 26, 2018 12:59 pm
- Forum: BioSemi Hardware
- Topic: Random '20202020...' triggers
- Replies: 6
- Views: 15684