external electrode with high impedance

Posted: Tue May 19, 2015 11:48 am
by mpwa
Hello,

One of our external electrodes has a constant high impedance. it is not the AD box, because if we use another external electrode, its impedance is fine. Is there anyway to fix this? the CM stays in range, that is the AD box has a solid blue light, so none of the other electrodes are affected.

Thanks
Martin

Re: external electrode with high impedance

Posted: Tue May 19, 2015 12:00 pm
by Coen
It's a high offset, not impedance.

If within +50 mV and - 50 mV and stable, then all is OK. If not, replace the electrode.

Best regards, Coen (BioSemi)