Hello,
One of our external electrodes has a constant high impedance. it is not the AD box, because if we use another external electrode, its impedance is fine. Is there anyway to fix this? the CM stays in range, that is the AD box has a solid blue light, so none of the other electrodes are affected.
Thanks
Martin
external electrode with high impedance
-
- Site Admin
- Posts: 1152
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: external electrode with high impedance
It's a high offset, not impedance.
If within +50 mV and - 50 mV and stable, then all is OK. If not, replace the electrode.
Best regards, Coen (BioSemi)