Hi,
I am working with a BioSemi Active 3 system with 64 channels with a GSR module. I tried recording EEG + GSR of three participants. For 1 out of those three I get 16Hz oscillations in all channels (see attached), for the other 2 I get 16Hz oscillations in 1 channel only (P1 and P2, respectively). In old posts I read that this 16HZ interference is normal with the DRL on the head, and that it can be removed with a sharp notch filter on 16Hz. Alternatively, it can be reduced by having the DRL between the hand and head. I have two questions I hope you might be able to help me with:
1. Do you have any idea as to why 2 out of my 3 participants show this 16Hz interference only in 1 channel, while the third participant has the 16Hz oscillations in every channel?
2. For an alternative DRL location (e.g. the neck) I suppose I would need a flat-DRL/pin-CMS. Is this currently available?
Thanks very much in advance!
Julie
GSR interference 16Hz oscillation
-
juliemichelle
- Posts: 1
- Joined: Wed May 27, 2026 10:37 am
GSR interference 16Hz oscillation
- Attachments
-
- 16Hzinterference_allch.png (235.31 KiB) Viewed 108 times
-
Coen
- Site Admin
- Posts: 1164
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: GSR interference 16Hz oscillation
It looks like a combination of a defect GSR sensor (most probably a short circuit to the isolated 16 Hz oscillator) and a gel bridge between DRL and P1/P2 (in the cases were only one channel shows 16 Hz). Please sent the GSR sensor to BioSemi for checking and replacement.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)