Error 5000

Post by irsikv »

hello,

I have been intermittently receiving the following error 5000 (The ADC box is off or the optical fiber is disconnected) when running ActiView. The blue light is illuminated on the box-- so I can confirm the battery is not the issue. Is it safe to assume that the optic fiber cord is going bad, or is it possible that the USB cord may be causing the error? Is there a straight forward way to confirm which one is the problem? (e.g., a meter).

Thanks,

Vanessa

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Error 5000

Post by Coen »

Common causes of intermittent 5000 errors are (ordered from most to least likely):

- bad optical fiber
- bad USB cable
- bad optical transmitter
- bad optical receiver

Replace fiber and USB cable. If the problem persists, then sent the AD-box and USB receiver to BioSemi for testing and repair.

Best regards, Coen (BioSemi)

Petter
Posts: 1
Joined: Thu Jan 28, 2016 1:47 pm

Re: Error 5000

Post by Petter »

Hi,
we have had a lot of error 5000 lately. I would suspect the USB-cable out of these options, can I change to any kind of USB-cable or are there some quality specifications that need to be met?

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Error 5000

Post by Coen »

Any USB cable is OK, also try a different USB port.

Best regards, Coen (BioSemi)