Only receiving odd (not even) triggers

Posted: Mon Jun 02, 2014 3:39 pm
by joshyouarebaker
I have created a program to send triggers 0 to 255. ActiView fails to detect '1', and then counts only odd numbers (it actually replaces even numbers with the previous odd number. Therefore, when sending triggers 0,1,2,3,4,5,6,7,8,9,10, ActiView displays 3,3,5,5,7,7,9,9.

any advice!?

Re: Only receiving odd (not even) triggers

Posted: Mon Jun 02, 2014 6:32 pm
by Coen
Line 1 always high (internal pullup), check cable, use analog trigger display (colored bars) for faultfinding.