Error 5000 occured at No Incoming Data

Post by jkronegg »

Dear all,

Offently, during data acquisition, I ActiView 5.21 software stop the data acquisition with the following error message:

Error 5000 occured at No Incoming Data
Possible reason(s):
ADC-box power is off
or
Optical fiber is disconnected

The battery green led (power) is on, the acquisition box "power" led and "CM in range" blue led are on continuously.

After closing the error dialog, I just click on the "Start" button to restart the acquisition system, so it doesn't seem that there is a problem with the optical fiber or the ADC-box...

Does anyone have the same error ?


Julien

Coen
Site Admin
Posts: 1125
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Post by Coen »

This is a known problem with some USB cards. A firmware update and Windows patch are available on the Biosemi websit, please see viewtopic.php?t=23

Best regards, Coen (BioSemi)

jkronegg
Posts: 10
Joined: Mon Jul 19, 2004 1:51 pm
Location: University of Geneva/CVML, Geneva, Switzerland
Contact:

Post by jkronegg »

Thank you. I just see it now while browsing the "hardware forum".

best regards,

Julien