Data stops recording

Post Reply
fbenjamin
Posts: 1
Joined: Tue Nov 12, 2019 8:11 pm

Data stops recording

Post by fbenjamin »

We are running version 4.09 and in the last three weeks started experiencing Actiview automatically changing from saving data to not saving data. It happens on about half of the participants. When we see the indicator that data is no longer being saved, we can select the button to start saving again and it starts saving the data and appends the data to the end of the current file. When looking at the data file, the location where the stop happened has a series of triggers(5-10) which are labeled in the 254-264 range. - Any ideas?

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Data stops recording

Post by Coen »

Your CFG is probably set to control the Pause button with triggers.

Edit the CFG in Notepad to (10th and 11th line from the end):

PauseOff=-1 //-1 is disabled, 0-255 is enabled
PauseOn=-1 //-1 is disabled, 0-255 is enabled

Best regards, Coen (BioSemi)

Post Reply