TCP tab in daisy-chaining

Post Reply
OO7
Posts: 2
Joined: Wed Jul 29, 2009 4:24 pm
Location: Enschede

TCP tab in daisy-chaining

Post by OO7 »

Hello,

When we are recording with single A/D box, ActiView displays the TCP tab as active so that it is possible to tell whether the server is on and when a client is connected. But when we try daisy-chaining with two boxes, TCP tab goes inactive as soon as the start button ıs pressed. We see the signals from Box1 and Box2, but nothing seems to be transmitted from the server. How is it possible to send data via TCP when using the daisy-chain mode?

We use 32 electrodes on each of our two BioSemi Mark II systems:
End Box, mode 0
2nd Box (fiber in and out), mode 1
Sampling rate: 512 (1/4)

Thanks in advance,
Hayrettin
Coen
Site Admin
Posts: 1166
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Post by Coen »

The current version of ActiView (6.05) does not support TCP/IP transfer in multibox mode. I will write some extra code to add this option. I will sent you a Beta version of the updated ActiView directly.

Best regards, Coen
OO7
Posts: 2
Joined: Wed Jul 29, 2009 4:24 pm
Location: Enschede

Post by OO7 »

Okay, thank you.
oferperl
Posts: 1
Joined: Tue Feb 14, 2017 8:25 am

Re: TCP tab in daisy-chaining

Post by oferperl »

Hi,
We just tackled the same issue recently - we can't access the TCP menu once daisy-chained mode is active.

Back in 2009 :D , you suggested some software workaround for this issue. I wonder where can I get it?

Thanks,

Best,

Ofer Perl
Coen
Site Admin
Posts: 1166
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: TCP tab in daisy-chaining

Post by Coen »

There will soon be a new version of ActiView, where the TCP/IP link is replaced by the option of direct access to the ringbuffer from remote computers over the network.

Best regards, Coen (BioSemi)
shehata
Posts: 1
Joined: Mon Jul 24, 2017 5:28 pm

Re: TCP tab in daisy-chaining

Post by shehata »

Hi,

We need to access the TCP menu in the daisy-chained mode. When the new version of ActiView will be available or can we get the unreleased ActiView or the beta version that can do this as with your answer to "007" back in 2009?

We are using 2 x MK2 A/D boxes.

Thanks,
M Shehata