Hi,
we're trying to use the Event related average module but without much success: even if the right speedmode is set, channels and trigger are correctly selected, no traces appear when pressing the 'Start Avg'. Could you please help me with this issue?
Thanks a lot in advance!
Event related average problem
-
Coen
- Site Admin
- Posts: 1163
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Is the acquisition running and do the triggers appear in the monopolar display (bottom of page) ?
Is the REJECTED SWEEPS counter counting ? If yes, then relax the conditions of the on-line artifact rejection. If no, then check the triggers settings (triggers are apparently not detected by the avg modules.
Best regards, Coen (BioSemi)
Is the REJECTED SWEEPS counter counting ? If yes, then relax the conditions of the on-line artifact rejection. If no, then check the triggers settings (triggers are apparently not detected by the avg modules.
Best regards, Coen (BioSemi)