CMS/DRL:basic question

Post Reply
laura
Posts: 2
Joined: Thu Sep 01, 2005 1:47 pm

CMS/DRL:basic question

Post by laura »

Hello!
This is a very basic question... :oops:
We are learning to use EEG 128 electrodes, and there is something which is unclear...Could anyone help us?
We've been told that Electrodes A +CMS/DRL have to be connected to the box before the other group of Electrodes (B, C and D). Is it right? If we don't proceed like that, will EEG recording be "wrong" ? And why?

Waiting for your aswers... :D

laura

Coen
Site Admin
Posts: 1125
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Post by Coen »

It does not matter in which order the electrodes connectors are plugged into the AD-box. However, valid EEG signals can only be measured with the CMS and DRL electrodes properly connected to the subject (blue light continuously on, see viewtopic.php?t=42)

Best regards, Coen (BioSemi)

laura
Posts: 2
Joined: Thu Sep 01, 2005 1:47 pm

Post by laura »

Thanks for your answer! :D