I have a very new P32-ABC-A cable. Today all a sudden, it is not working in channel A, only in channel B. Does anybody have similar experience?
-
- Site Admin
- Posts: 1125
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
The A set should work in the B receptacle (and in all other receptacles that are equipped with modules), and the B set (and all other sets) should work in the A receptacle (provided a separate CMS/DRL is used). Set up an one bucket water test, and locate the problem (set or receptacle) by swapping connectors and sets. Then sent the defect parts to your BioSemi dealer for repair or replacement.
Be careful to inspect the connectors before the test. If any pins are bent, then return the set for repair. Forcing a connector with bent pins into a receptacle will inevitably damage the (costly) receptacle.
Best regards, Coen (BioSemi)