Posted: Thu Dec 02, 2021 4:03 pm
by JonasW
Hey,
We are currently planning to use the Galvanic Skin Response by Biosemi in upcoming experiments. We have an ActiveTwo AD-box, but are not sure if it is possible to use GSR with our version of the box.

The box was ordered a long time ago and the responsible people are no longer working in our group: Our box does not have the GSR label on any of the aux inputs (the labels are "CMS/DRL", "Ergo1" and "aux 3").

Am I correct in assuming that the GSR would not be compatible with our box or is it possible to use the generic input ("aux 3")?

Thank you very much!

Re: Finding GSR input

Posted: Fri Dec 03, 2021 3:07 pm
by Coen
If there is no GSR label, then there is no GSR input. Aux3 (unlabeled) is currently not connected. BioSemi can install a GSR input on Aux3.

Best regards, Coen (BioSemi)