Finding GSR input

Post Reply
JonasW
Posts: 1
Joined: Thu Dec 02, 2021 3:53 pm

Finding GSR input

Post by JonasW »

Hey,
We are currently planning to use the Galvanic Skin Response by Biosemi in upcoming experiments. We have an ActiveTwo AD-box, but are not sure if it is possible to use GSR with our version of the box.

The box was ordered a long time ago and the responsible people are no longer working in our group: Our box does not have the GSR label on any of the aux inputs (the labels are "CMS/DRL", "Ergo1" and "aux 3").

Am I correct in assuming that the GSR would not be compatible with our box or is it possible to use the generic input ("aux 3")?

Thank you very much!

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Finding GSR input

Post by Coen »

If there is no GSR label, then there is no GSR input. Aux3 (unlabeled) is currently not connected. BioSemi can install a GSR input on Aux3.

Best regards, Coen (BioSemi)