Search found 3 matches

by bisbel
Fri Mar 22, 2019 1:03 pm
Forum: BioSemi Hardware
Topic: Fault with ActiveTwo AD Box
Replies: 4
Views: 11640

Re: Fault with ActiveTwo AD Box

Thanks Coen, Just a question. Your website says that when such a gel bridge occurs in the connectors on the AD box, then the AD box will go into shutdown without any electrodes connected (https://www.biosemi.com/faq/AD-box%20shuts-down%20or%20Acquisition%20stops.htm). However, this is not the case w...
by bisbel
Fri Mar 22, 2019 7:53 am
Forum: BioSemi Hardware
Topic: Fault with ActiveTwo AD Box
Replies: 4
Views: 11640

Re: Fault with ActiveTwo AD Box

In addition, I have checked the electrode connector housings for gel contamination in case of short circuit/bridging was the fault. However, all connector ports and pins appear clean and free from gel or other contamination.

Sincerely,
Ben Isbel
by bisbel
Fri Mar 22, 2019 7:49 am
Forum: BioSemi Hardware
Topic: Fault with ActiveTwo AD Box
Replies: 4
Views: 11640

Fault with ActiveTwo AD Box

Hi Coen, I have a problem that I hope you can help with. We currently have a problem with our AD box automatically powering off when the CMS/DRL circuit is complete. To troubleshoot this problem, I did a bucket test with only the CMS/DRL aux electrodes with the following results. Without any electro...