Search found 1158 matches

by Coen
Wed Nov 12, 2025 7:08 pm
Forum: BioSemi Software (ActiView executable)
Topic: ActiView103 sending multiple LSL streams
Replies: 5
Views: 5558

Re: ActiView103 sending multiple LSL streams

I have made some changes to the LabVIEW code that should prevent the multiple streams. Please email to info@biosemi.com and I will sent you a download link for a preliminary version of the upcoming ActiView 104.

Best regards, Coen (BioSemi)
by Coen
Wed Nov 12, 2025 3:31 pm
Forum: EEG data analysis
Topic: Gain and offset for physical values
Replies: 7
Views: 43988

Re: Gain and offset for physical values

BDF files by A2 and A3 are basically the same, you can ignore the offset parameter. Scaling for the EEG channels is the same (LSB = 31.25 nV = 1/32 uV).

A difference between A2 and A3 BDF files is the scaling of some of the auxilary sensors, as defined in the BDF header.

In addition, the analog ...
by Coen
Mon Nov 10, 2025 5:47 pm
Forum: BioSemi all other topics
Topic: Conductive floor
Replies: 1
Views: 2579

Re: Conductive floor

The ESD floor has a high resistance (MegaOhm to GigaOhm) and would not make any difference to the interference rejection of the BioSemi system.

The inputs of the BioSemi system are protected against static discharge (standard EN 61326). An ESD floor is not required to prevent damage. Just remain ...
by Coen
Fri Oct 31, 2025 1:11 pm
Forum: BioSemi Software (ActiView executable)
Topic: ActiView103 sending multiple LSL streams
Replies: 5
Views: 5558

Re: ActiView103 sending multiple LSL streams

Does it work with the BrainVision LSL viewer ( https://labstreaminglayer.readthedocs.i ... ewers.html ) ?

Which errors do you see in ActiView (popup messages and/or error counters in the lower right corner) ?

Best regards, Coen (BioSemi)
by Coen
Mon Jun 16, 2025 9:28 pm
Forum: BioSemi all other topics
Topic: Real-time processing of EEG signals from Active3
Replies: 3
Views: 112794

Re: Real-time processing of EEG signals from Active3

https://www.linkedin.com/posts/neurospec_news-from-the-house-of-biosemi-with-the-activity-7340398038353416193-BpTW?utm_source=share&utm_medium=member_desktop&rcm=ACoAAAKVmwQBf6tXt0CAtOT6Bv7Kx3t821cL4Pw

ActiView 10.2 for ActievThree with native LSL support will be on the BioSemi download page soon ...
by Coen
Mon Jun 09, 2025 1:52 pm
Forum: BioSemi Hardware
Topic: Simultaneous EEG + EMG
Replies: 1
Views: 21019

Re: Simultaneous EEG + EMG

You should only connect one CMS/DRL to the AD-box (otherwise, the blue light keeps flashing). Connect either the A set with CMS/DRL, or the front-panel CMS/DRL. So, you can use the EX electrodes with the front-panel CMS/DRL provided the A connector remains unplugged.

Best regards, Coen (BioSemi)
by Coen
Wed Apr 02, 2025 1:39 pm
Forum: BioSemi Software (ActiView executable)
Topic: Electrode labels when checking offsets in ActiView
Replies: 1
Views: 45245

Re: Electrode labels when checking offsets in ActiView

Use the CFG file for 64 channels with 1020 labeling, best regards, Coen (BioSemi)
by Coen
Tue Feb 11, 2025 10:21 pm
Forum: BioSemi Hardware
Topic: Updating firmware for high speed
Replies: 2
Views: 72049

Re: Updating firmware for high speed

Contact info@biosemi.com

You will have to ship your AD-box and USB-receiver to BioSemi.

Note that the upgrade is only possible for AD-box serial ADC16 or higher.

Best regards, Coen (BioSemi)
by Coen
Tue Feb 04, 2025 3:52 pm
Forum: BioSemi Hardware
Topic: Need Advise!!
Replies: 1
Views: 24265

Re: Need Advise!!

Since the skin is not abraded, disinfection is generally not needed. When disinfection is preferred (as was the case during the Covid-19 outbreak), then follow the instructions as shown in https://www.youtube.com/watch?v=-CZx3_tbrYM . Do not use hydrogen peroxide since it may damage the headcap ...
by Coen
Fri Nov 15, 2024 7:45 pm
Forum: BioSemi Hardware
Topic: Error 5000
Replies: 13
Views: 106150

Re: Error 5000

Contact your local BioSemi representative
by Coen
Fri Nov 15, 2024 6:36 pm
Forum: BioSemi Hardware
Topic: Error 5000
Replies: 13
Views: 106150

Re: Error 5000

ADC6 is an Mk1 AD-Box and does not work with the red receiver. Your white receiver does work with the USB Trigger Interface (the trigger ports of white and red receivers are identical).

Best regards, Coen (BioSemi)
by Coen
Fri Nov 15, 2024 1:44 pm
Forum: BioSemi all other topics
Topic: Real-time processing of EEG signals from Active3
Replies: 3
Views: 112794

Re: Real-time processing of EEG signals from Active3

ActiView for ActiveThree can stream the data from the ringbuffer via a TCP/IP connection. You can consider to develop a MatLab client to read this real-time data stream.

Email to info@biosemi.com for the latest ActiView version and a TCP/IP client example.

Best regards, Coen (BioSemi)
by Coen
Tue Nov 12, 2024 5:59 pm
Forum: BioSemi Software (ActiView executable)
Topic: ActiView10 not working with USB2 receiver with Firmware4.0
Replies: 4
Views: 43251

Re: ActiView10 not working with USB2 receiver with Firmware4.0

It is normal operation that only trigger lines 15 (start/stop) and 16 (1 sec pulses) are transmitted from Master to Slave. The stimulation triggers should indeed be connected to the Master. The trigger lines 1-14 of the Slave are ignored when the Master and Slave files are merged into a single ...
by Coen
Mon Nov 11, 2024 3:01 pm
Forum: BioSemi Software (ActiView executable)
Topic: ActiView10 not working with USB2 receiver with Firmware4.0
Replies: 4
Views: 43251

Re: ActiView10 not working with USB2 receiver with Firmware4.0

ActiveTwo (Mk2 and Mk2.5) only works with ActiView 9.xx

ActiveThree only works with ActiView 10.x

Please contact your local BioSemi represenative for ActiView 9.03 that allows hypersampling:

Master: ActiveThree with ActiView 10.0 on 16 kHz sample rate
Slave: ActiveTwo with ActiView 9.03 on 2 kHz ...
by Coen
Wed Nov 06, 2024 6:30 pm
Forum: BioSemi Software (ActiView executable)
Topic: Errors 7000 and 7001
Replies: 2
Views: 29514

Re: Errors 7000 and 7001

The above CFG is for ActiveTwo.

Please download the ActiveThree CFGs from https://www.biosemi.com/download/CFG_fi ... view10.zip

Best regards, Coen (BioSemi)
by Coen
Thu Oct 31, 2024 3:41 pm
Forum: BioSemi Hardware
Topic: hardware filters
Replies: 3
Views: 36347

Re: hardware filters

Which "analog box" exactly are you talking about ?

It must be something else than a BioSemi AIB (https://www.biosemi.com/aib.htm) because the BioSemi box only has a digital optical output (and it is unlikely that you can measure anything meaningful from this optical data stream).
by Coen
Thu Oct 31, 2024 3:25 pm
Forum: BioSemi all other topics
Topic: BIOSEMI Trigger and EEGLAB reading events
Replies: 1
Views: 31090

Re: BIOSEMI Trigger and EEGLAB reading events

ActiveTwo Mk1 and Mk2 (as well as the new ActiveThree) systems use exactly the same BDF file format (https://www.biosemi.com/faq/file_format.htm). In particular, storing of the 16 trigger lines in the Status channel is the same for all systems. The only difference between Mk1 and Mk2 files is the ...
by Coen
Thu Oct 24, 2024 6:38 pm
Forum: BioSemi Software (ActiView executable)
Topic: Older Versions Repository? Looking for ActiView Version 705
Replies: 1
Views: 25703

Re: Older Versions Repository? Looking for ActiView Version 705

Use the current 9.02 version, it works with every ActiveTwo system (even as old as from 2002). You might have to upgrade the firmware of your USB-receiver to use it with Windows 10/11, see https://www.biosemi.com/faq/install_USB.htm.

Best regards, Coen (BioSemi)
by Coen
Wed Oct 02, 2024 1:52 pm
Forum: BioSemi Hardware
Topic: hardware filters
Replies: 3
Views: 36347

Re: hardware filters

The BioSemi system has DC amplifiers, there is no high-pass filter. Check the signal in ActiView with the (display) filters switched off. You have either a high-pass in your file reader, or you have capacitively coupled the trigger line to the Analog Input Box.

Best regards, Coen (BioSemi)
by Coen
Wed Sep 25, 2024 1:32 pm
Forum: BioSemi all other topics
Topic: Hyperscanning with LabStreamingLayer
Replies: 6
Views: 50447

Re: Hyperscanning with LabStreamingLayer

With the triggered hyperscanning method, the data files from the two (or more) systems are merged into a single hyperscanning file offline. So it is not possible to stream hyperscanning data during the recording.

Best regards, Coen (BioSemi)
by Coen
Tue Sep 24, 2024 3:19 pm
Forum: BioSemi all other topics
Topic: Hyperscanning with LabStreamingLayer
Replies: 6
Views: 50447

Re: Hyperscanning with LabStreamingLayer

According to the LSL website it should work with ActiveTwo (2 and 2.5), but that's all I know.

Best regards, Coen (BioSemi)
by Coen
Mon Sep 23, 2024 2:30 pm
Forum: BioSemi all other topics
Topic: Hyperscanning with LabStreamingLayer
Replies: 6
Views: 50447

Re: Hyperscanning with LabStreamingLayer

You can Hyperscan with the Active3 and Active2.5 system using the method described on https://www.biosemi.com/faq/Hyperscanni ... erface.htm .

It is not possible to use LSL for ActiveThree Hyperscanning.

Best regards, Coen (BioSemi)
by Coen
Mon Sep 09, 2024 6:24 pm
Forum: BioSemi Hardware
Topic: Error 5000
Replies: 13
Views: 106150

Re: Error 5000

Send the red receiver to BioSemi, we will replace it with a white one (we still have a few).

Best regards, Coen (BioSemi)
by Coen
Fri Sep 06, 2024 5:24 pm
Forum: BioSemi Hardware
Topic: Error 5000
Replies: 13
Views: 106150

Re: Error 5000

ADC11 are Mk1 systems (20 years old), they do not work with the new red receiver. You should use the older type white receiver.

Best regards, Coen (BioSemi)
by Coen
Fri Sep 06, 2024 4:15 pm
Forum: BioSemi Hardware
Topic: Error 5000
Replies: 13
Views: 106150

Re: Error 5000

What is the serial number of your AD-box ?

Best regards, Coen (BioSemi)
by Coen
Tue Aug 06, 2024 2:33 pm
Forum: BioSemi Software (ActiView executable)
Topic: How to open .cgf
Replies: 1
Views: 30849

Re: How to open .cgf

The CFG file stores all ActiView control settings, the channel labels, and a few system settings (e.g. how response switch signals are linked to triggers). ActiView opens with the default.cfg file.

On the About/Configure tabpage, you can read a CFG file (for example a CFG with 32-chan settings and ...
by Coen
Mon Jul 15, 2024 6:03 pm
Forum: BioSemi Software (ActiView executable)
Topic: Detailed review on the concept of offset?
Replies: 4
Views: 45933

Re: Detailed review on the concept of offset?

Google on "opamp buffer output impedance". With active electrodes, the electrode impedances in the interference paper (Ze in Fig. 1) are reduced from tens of kOhm to less than an Ohm, and the magnitude of interference is reduced with the same ratio.

Best regards, Coen (BioSemi)