I have made some changes to the LabVIEW code that should prevent the multiple streams. Please email to info@biosemi.com and I will sent you a download link for a preliminary version of the upcoming ActiView 104.
Best regards, Coen (BioSemi)
Search found 1158 matches
- Wed Nov 12, 2025 7:08 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: ActiView103 sending multiple LSL streams
- Replies: 5
- Views: 5558
- Wed Nov 12, 2025 3:31 pm
- Forum: EEG data analysis
- Topic: Gain and offset for physical values
- Replies: 7
- Views: 43988
Re: Gain and offset for physical values
BDF files by A2 and A3 are basically the same, you can ignore the offset parameter. Scaling for the EEG channels is the same (LSB = 31.25 nV = 1/32 uV).
A difference between A2 and A3 BDF files is the scaling of some of the auxilary sensors, as defined in the BDF header.
In addition, the analog ...
A difference between A2 and A3 BDF files is the scaling of some of the auxilary sensors, as defined in the BDF header.
In addition, the analog ...
- Mon Nov 10, 2025 5:47 pm
- Forum: BioSemi all other topics
- Topic: Conductive floor
- Replies: 1
- Views: 2579
Re: Conductive floor
The ESD floor has a high resistance (MegaOhm to GigaOhm) and would not make any difference to the interference rejection of the BioSemi system.
The inputs of the BioSemi system are protected against static discharge (standard EN 61326). An ESD floor is not required to prevent damage. Just remain ...
The inputs of the BioSemi system are protected against static discharge (standard EN 61326). An ESD floor is not required to prevent damage. Just remain ...
- Fri Oct 31, 2025 1:11 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: ActiView103 sending multiple LSL streams
- Replies: 5
- Views: 5558
Re: ActiView103 sending multiple LSL streams
Does it work with the BrainVision LSL viewer ( https://labstreaminglayer.readthedocs.i ... ewers.html ) ?
Which errors do you see in ActiView (popup messages and/or error counters in the lower right corner) ?
Best regards, Coen (BioSemi)
Which errors do you see in ActiView (popup messages and/or error counters in the lower right corner) ?
Best regards, Coen (BioSemi)
- Mon Jun 16, 2025 9:28 pm
- Forum: BioSemi all other topics
- Topic: Real-time processing of EEG signals from Active3
- Replies: 3
- Views: 112794
Re: Real-time processing of EEG signals from Active3
https://www.linkedin.com/posts/neurospec_news-from-the-house-of-biosemi-with-the-activity-7340398038353416193-BpTW?utm_source=share&utm_medium=member_desktop&rcm=ACoAAAKVmwQBf6tXt0CAtOT6Bv7Kx3t821cL4Pw
ActiView 10.2 for ActievThree with native LSL support will be on the BioSemi download page soon ...
ActiView 10.2 for ActievThree with native LSL support will be on the BioSemi download page soon ...
- Mon Jun 09, 2025 1:52 pm
- Forum: BioSemi Hardware
- Topic: Simultaneous EEG + EMG
- Replies: 1
- Views: 21019
Re: Simultaneous EEG + EMG
You should only connect one CMS/DRL to the AD-box (otherwise, the blue light keeps flashing). Connect either the A set with CMS/DRL, or the front-panel CMS/DRL. So, you can use the EX electrodes with the front-panel CMS/DRL provided the A connector remains unplugged.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Wed Apr 02, 2025 1:39 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Electrode labels when checking offsets in ActiView
- Replies: 1
- Views: 45245
Re: Electrode labels when checking offsets in ActiView
Use the CFG file for 64 channels with 1020 labeling, best regards, Coen (BioSemi)
- Tue Feb 11, 2025 10:21 pm
- Forum: BioSemi Hardware
- Topic: Updating firmware for high speed
- Replies: 2
- Views: 72049
Re: Updating firmware for high speed
Contact info@biosemi.com
You will have to ship your AD-box and USB-receiver to BioSemi.
Note that the upgrade is only possible for AD-box serial ADC16 or higher.
Best regards, Coen (BioSemi)
You will have to ship your AD-box and USB-receiver to BioSemi.
Note that the upgrade is only possible for AD-box serial ADC16 or higher.
Best regards, Coen (BioSemi)
- Tue Feb 04, 2025 3:52 pm
- Forum: BioSemi Hardware
- Topic: Need Advise!!
- Replies: 1
- Views: 24265
Re: Need Advise!!
Since the skin is not abraded, disinfection is generally not needed. When disinfection is preferred (as was the case during the Covid-19 outbreak), then follow the instructions as shown in https://www.youtube.com/watch?v=-CZx3_tbrYM . Do not use hydrogen peroxide since it may damage the headcap ...
- Tue Feb 04, 2025 3:35 pm
- Forum: BioSemi Hardware
- Topic: Compatibility with Transcutaneous vagus nerve stimulation (tVNS)
- Replies: 1
- Views: 24950
- Tue Dec 17, 2024 7:50 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Access Offsets from.bdf files
- Replies: 1
- Views: 58741
- Fri Nov 15, 2024 7:45 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 106150
Re: Error 5000
Contact your local BioSemi representative
- Fri Nov 15, 2024 6:36 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 106150
Re: Error 5000
ADC6 is an Mk1 AD-Box and does not work with the red receiver. Your white receiver does work with the USB Trigger Interface (the trigger ports of white and red receivers are identical).
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Nov 15, 2024 1:44 pm
- Forum: BioSemi all other topics
- Topic: Real-time processing of EEG signals from Active3
- Replies: 3
- Views: 112794
Re: Real-time processing of EEG signals from Active3
ActiView for ActiveThree can stream the data from the ringbuffer via a TCP/IP connection. You can consider to develop a MatLab client to read this real-time data stream.
Email to info@biosemi.com for the latest ActiView version and a TCP/IP client example.
Best regards, Coen (BioSemi)
Email to info@biosemi.com for the latest ActiView version and a TCP/IP client example.
Best regards, Coen (BioSemi)
- Tue Nov 12, 2024 5:59 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: ActiView10 not working with USB2 receiver with Firmware4.0
- Replies: 4
- Views: 43251
Re: ActiView10 not working with USB2 receiver with Firmware4.0
It is normal operation that only trigger lines 15 (start/stop) and 16 (1 sec pulses) are transmitted from Master to Slave. The stimulation triggers should indeed be connected to the Master. The trigger lines 1-14 of the Slave are ignored when the Master and Slave files are merged into a single ...
- Mon Nov 11, 2024 3:01 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: ActiView10 not working with USB2 receiver with Firmware4.0
- Replies: 4
- Views: 43251
Re: ActiView10 not working with USB2 receiver with Firmware4.0
ActiveTwo (Mk2 and Mk2.5) only works with ActiView 9.xx
ActiveThree only works with ActiView 10.x
Please contact your local BioSemi represenative for ActiView 9.03 that allows hypersampling:
Master: ActiveThree with ActiView 10.0 on 16 kHz sample rate
Slave: ActiveTwo with ActiView 9.03 on 2 kHz ...
ActiveThree only works with ActiView 10.x
Please contact your local BioSemi represenative for ActiView 9.03 that allows hypersampling:
Master: ActiveThree with ActiView 10.0 on 16 kHz sample rate
Slave: ActiveTwo with ActiView 9.03 on 2 kHz ...
- Wed Nov 06, 2024 6:30 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Errors 7000 and 7001
- Replies: 2
- Views: 29514
Re: Errors 7000 and 7001
The above CFG is for ActiveTwo.
Please download the ActiveThree CFGs from https://www.biosemi.com/download/CFG_fi ... view10.zip
Best regards, Coen (BioSemi)
Please download the ActiveThree CFGs from https://www.biosemi.com/download/CFG_fi ... view10.zip
Best regards, Coen (BioSemi)
- Thu Oct 31, 2024 3:41 pm
- Forum: BioSemi Hardware
- Topic: hardware filters
- Replies: 3
- Views: 36347
Re: hardware filters
Which "analog box" exactly are you talking about ?
It must be something else than a BioSemi AIB (https://www.biosemi.com/aib.htm) because the BioSemi box only has a digital optical output (and it is unlikely that you can measure anything meaningful from this optical data stream).
It must be something else than a BioSemi AIB (https://www.biosemi.com/aib.htm) because the BioSemi box only has a digital optical output (and it is unlikely that you can measure anything meaningful from this optical data stream).
- Thu Oct 31, 2024 3:25 pm
- Forum: BioSemi all other topics
- Topic: BIOSEMI Trigger and EEGLAB reading events
- Replies: 1
- Views: 31090
Re: BIOSEMI Trigger and EEGLAB reading events
ActiveTwo Mk1 and Mk2 (as well as the new ActiveThree) systems use exactly the same BDF file format (https://www.biosemi.com/faq/file_format.htm). In particular, storing of the 16 trigger lines in the Status channel is the same for all systems. The only difference between Mk1 and Mk2 files is the ...
- Thu Oct 24, 2024 6:38 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Older Versions Repository? Looking for ActiView Version 705
- Replies: 1
- Views: 25703
Re: Older Versions Repository? Looking for ActiView Version 705
Use the current 9.02 version, it works with every ActiveTwo system (even as old as from 2002). You might have to upgrade the firmware of your USB-receiver to use it with Windows 10/11, see https://www.biosemi.com/faq/install_USB.htm.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Wed Oct 02, 2024 1:52 pm
- Forum: BioSemi Hardware
- Topic: hardware filters
- Replies: 3
- Views: 36347
Re: hardware filters
The BioSemi system has DC amplifiers, there is no high-pass filter. Check the signal in ActiView with the (display) filters switched off. You have either a high-pass in your file reader, or you have capacitively coupled the trigger line to the Analog Input Box.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Wed Sep 25, 2024 1:32 pm
- Forum: BioSemi all other topics
- Topic: Hyperscanning with LabStreamingLayer
- Replies: 6
- Views: 50447
Re: Hyperscanning with LabStreamingLayer
With the triggered hyperscanning method, the data files from the two (or more) systems are merged into a single hyperscanning file offline. So it is not possible to stream hyperscanning data during the recording.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Tue Sep 24, 2024 3:19 pm
- Forum: BioSemi all other topics
- Topic: Hyperscanning with LabStreamingLayer
- Replies: 6
- Views: 50447
Re: Hyperscanning with LabStreamingLayer
According to the LSL website it should work with ActiveTwo (2 and 2.5), but that's all I know.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Mon Sep 23, 2024 2:30 pm
- Forum: BioSemi all other topics
- Topic: Hyperscanning with LabStreamingLayer
- Replies: 6
- Views: 50447
Re: Hyperscanning with LabStreamingLayer
You can Hyperscan with the Active3 and Active2.5 system using the method described on https://www.biosemi.com/faq/Hyperscanni ... erface.htm .
It is not possible to use LSL for ActiveThree Hyperscanning.
Best regards, Coen (BioSemi)
It is not possible to use LSL for ActiveThree Hyperscanning.
Best regards, Coen (BioSemi)
- Mon Sep 09, 2024 6:24 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 106150
Re: Error 5000
Send the red receiver to BioSemi, we will replace it with a white one (we still have a few).
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Sep 06, 2024 5:24 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 106150
Re: Error 5000
ADC11 are Mk1 systems (20 years old), they do not work with the new red receiver. You should use the older type white receiver.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Fri Sep 06, 2024 4:15 pm
- Forum: BioSemi Hardware
- Topic: Error 5000
- Replies: 13
- Views: 106150
Re: Error 5000
What is the serial number of your AD-box ?
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
- Tue Aug 06, 2024 2:33 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: How to open .cgf
- Replies: 1
- Views: 30849
Re: How to open .cgf
The CFG file stores all ActiView control settings, the channel labels, and a few system settings (e.g. how response switch signals are linked to triggers). ActiView opens with the default.cfg file.
On the About/Configure tabpage, you can read a CFG file (for example a CFG with 32-chan settings and ...
On the About/Configure tabpage, you can read a CFG file (for example a CFG with 32-chan settings and ...
- Tue Aug 06, 2024 2:03 pm
- Forum: BioSemi Hardware
- Topic: CMS/DRL are unstable
- Replies: 1
- Views: 33207
- Mon Jul 15, 2024 6:03 pm
- Forum: BioSemi Software (ActiView executable)
- Topic: Detailed review on the concept of offset?
- Replies: 4
- Views: 45933
Re: Detailed review on the concept of offset?
Google on "opamp buffer output impedance". With active electrodes, the electrode impedances in the interference paper (Ze in Fig. 1) are reduced from tens of kOhm to less than an Ohm, and the magnitude of interference is reduced with the same ratio.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)