Search found 1142 matches

by Coen
Wed Nov 06, 2024 6:30 pm
Forum: BioSemi Software (ActiView executable)
Topic: Errors 7000 and 7001
Replies: 2
Views: 61

Re: Errors 7000 and 7001

The above CFG is for ActiveTwo.

Please download the ActiveThree CFGs from https://www.biosemi.com/download/CFG_fi ... view10.zip

Best regards, Coen (BioSemi)
by Coen
Thu Oct 31, 2024 3:41 pm
Forum: BioSemi Hardware
Topic: hardware filters
Replies: 3
Views: 1654

Re: hardware filters

Which "analog box" exactly are you talking about ?

It must be something else than a BioSemi AIB (https://www.biosemi.com/aib.htm) because the BioSemi box only has a digital optical output (and it is unlikely that you can measure anything meaningful from this optical data stream).
by Coen
Thu Oct 31, 2024 3:25 pm
Forum: BioSemi all other topics
Topic: BIOSEMI Trigger and EEGLAB reading events
Replies: 1
Views: 174

Re: BIOSEMI Trigger and EEGLAB reading events

ActiveTwo Mk1 and Mk2 (as well as the new ActiveThree) systems use exactly the same BDF file format (https://www.biosemi.com/faq/file_format.htm). In particular, storing of the 16 trigger lines in the Status channel is the same for all systems. The only difference between Mk1 and Mk2 files is the MS...
by Coen
Thu Oct 24, 2024 6:38 pm
Forum: BioSemi Software (ActiView executable)
Topic: Older Versions Repository? Looking for ActiView Version 705
Replies: 1
Views: 410

Re: Older Versions Repository? Looking for ActiView Version 705

Use the current 9.02 version, it works with every ActiveTwo system (even as old as from 2002). You might have to upgrade the firmware of your USB-receiver to use it with Windows 10/11, see https://www.biosemi.com/faq/install_USB.htm.

Best regards, Coen (BioSemi)
by Coen
Wed Oct 02, 2024 1:52 pm
Forum: BioSemi Hardware
Topic: hardware filters
Replies: 3
Views: 1654

Re: hardware filters

The BioSemi system has DC amplifiers, there is no high-pass filter. Check the signal in ActiView with the (display) filters switched off. You have either a high-pass in your file reader, or you have capacitively coupled the trigger line to the Analog Input Box.

Best regards, Coen (BioSemi)
by Coen
Wed Sep 25, 2024 1:32 pm
Forum: BioSemi all other topics
Topic: Hyperscanning with LabStreamingLayer
Replies: 6
Views: 3376

Re: Hyperscanning with LabStreamingLayer

With the triggered hyperscanning method, the data files from the two (or more) systems are merged into a single hyperscanning file offline. So it is not possible to stream hyperscanning data during the recording.

Best regards, Coen (BioSemi)
by Coen
Tue Sep 24, 2024 3:19 pm
Forum: BioSemi all other topics
Topic: Hyperscanning with LabStreamingLayer
Replies: 6
Views: 3376

Re: Hyperscanning with LabStreamingLayer

According to the LSL website it should work with ActiveTwo (2 and 2.5), but that's all I know.

Best regards, Coen (BioSemi)
by Coen
Mon Sep 23, 2024 2:30 pm
Forum: BioSemi all other topics
Topic: Hyperscanning with LabStreamingLayer
Replies: 6
Views: 3376

Re: Hyperscanning with LabStreamingLayer

You can Hyperscan with the Active3 and Active2.5 system using the method described on https://www.biosemi.com/faq/Hyperscanni ... erface.htm .

It is not possible to use LSL for ActiveThree Hyperscanning.

Best regards, Coen (BioSemi)
by Coen
Mon Sep 09, 2024 6:24 pm
Forum: BioSemi Hardware
Topic: Error 5000
Replies: 9
Views: 16933

Re: Error 5000

Send the red receiver to BioSemi, we will replace it with a white one (we still have a few).

Best regards, Coen (BioSemi)
by Coen
Fri Sep 06, 2024 5:24 pm
Forum: BioSemi Hardware
Topic: Error 5000
Replies: 9
Views: 16933

Re: Error 5000

ADC11 are Mk1 systems (20 years old), they do not work with the new red receiver. You should use the older type white receiver.

Best regards, Coen (BioSemi)
by Coen
Fri Sep 06, 2024 4:15 pm
Forum: BioSemi Hardware
Topic: Error 5000
Replies: 9
Views: 16933

Re: Error 5000

What is the serial number of your AD-box ?

Best regards, Coen (BioSemi)
by Coen
Tue Aug 06, 2024 2:33 pm
Forum: BioSemi Software (ActiView executable)
Topic: How to open .cgf
Replies: 1
Views: 3138

Re: How to open .cgf

The CFG file stores all ActiView control settings, the channel labels, and a few system settings (e.g. how response switch signals are linked to triggers). ActiView opens with the default.cfg file. On the About/Configure tabpage, you can read a CFG file (for example a CFG with 32-chan settings and 1...
by Coen
Mon Jul 15, 2024 6:03 pm
Forum: BioSemi Software (ActiView executable)
Topic: Detailed review on the concept of offset?
Replies: 4
Views: 5531

Re: Detailed review on the concept of offset?

Google on "opamp buffer output impedance". With active electrodes, the electrode impedances in the interference paper (Ze in Fig. 1) are reduced from tens of kOhm to less than an Ohm, and the magnitude of interference is reduced with the same ratio.

Best regards, Coen (BioSemi)
by Coen
Fri Jul 12, 2024 7:15 pm
Forum: BioSemi Software (ActiView executable)
Topic: Detailed review on the concept of offset?
Replies: 4
Views: 5531

Re: Detailed review on the concept of offset?

1. What is the definition of offset? The electrode-gel-skin interface forms a galvanic half-cell. The half-cell potential is determined by the electrode metal. The offset voltage is the potential difference between two of these half-cells. When the electrodes are identical (same metal and perfectly ...
by Coen
Fri Jun 07, 2024 2:36 pm
Forum: BioSemi all other topics
Topic: Modify ActiView Configuration
Replies: 1
Views: 9086

Re: Modify ActiView Configuration

The above is not a standard option in ActiView that can be enabled with a CFG setting. Bipolar display for all channels would require changing the (open source LabVIEW) source code.

Best regards, Coen (BioSemi)
by Coen
Thu May 30, 2024 1:38 pm
Forum: EEG data analysis
Topic: Seeking help with CMS and DRL location with EEGLAB
Replies: 2
Views: 12112

Re: Seeking help with CMS and DRL location with EEGLAB

Reference the data. After referencing, the locations of CMS and DRL are not relevant anymore.

Best regards, Coen (BioSemi)
by Coen
Sat Mar 09, 2024 5:48 pm
Forum: BioSemi Software (ActiView executable)
Topic: Event Related Average module
Replies: 4
Views: 14429

Re: Event Related Average module

That depends on whether the evoked responses also alternate in polarity. With ABR for example, that's not the case (so the noise trace still shows the noise).

Best regards, Coen (BioSemi)
by Coen
Fri Mar 08, 2024 7:23 pm
Forum: BioSemi Hardware
Topic: Active Two Questions (CM. ABR, etc.)
Replies: 4
Views: 10036

Re: Active Two Questions (CM. ABR, etc.)

The trigger port needs TTL level signals. In order to connect an audio line signal, you will need to convert to TTL with a comparator and flip-flop. The accuracy of trigger timing is one sample length, see https://www.biosemi.com/faq/review_trigger_timing.htm

Best regards, Coen
by Coen
Fri Mar 08, 2024 1:56 pm
Forum: BioSemi Hardware
Topic: Active Two Questions (CM. ABR, etc.)
Replies: 4
Views: 10036

Re: Active Two Questions (CM. ABR, etc.)

It is a custom made system for 16+3 channel ECG (LA RA LL is the Wilson Central Terminal). It is probably something like 20 years old. Please email the serial number to info@biosemi.com, and I can check which upgrades are still possible. Soundcard to trigger interface can be done with the ERGO-OPTIC...
by Coen
Wed Feb 14, 2024 2:42 pm
Forum: BioSemi Software (ActiView executable)
Topic: Editing 32-channels in default.cfg
Replies: 3
Views: 7696

Re: Editing 32-channels in default.cfg

Do not edit CFGs in Notepad.

Open ActiView, set the controls according to your preferences, make a short data file, select the "about" tabpage and save your front-panel settings in a new CFG.

Best regards, Coen (BioSemi)
by Coen
Wed Feb 14, 2024 2:37 pm
Forum: BioSemi all other topics
Topic: Use few electrodes than64
Replies: 2
Views: 8284

Re: Use few electrodes than64

No, best regards, Coen (BioSemi)
by Coen
Tue Nov 07, 2023 2:47 pm
Forum: BioSemi Hardware
Topic: Biosemi simultaneous Biopac
Replies: 1
Views: 8723

Re: Biosemi simultaneous Biopac

Why don't you measure simultaneous EEG and ECG (extra TP-flat electrodes) with the BioSemi system ?

Best regards, Coen (BioSemi)
by Coen
Tue Oct 31, 2023 3:11 pm
Forum: EEG data analysis
Topic: How Can I Verify That ActiveTwo USB-Receiver 5.1 Is Receiving Triggers from PsychoPy 3?
Replies: 4
Views: 12731

Re: How Can I Verify That ActiveTwo USB-Receiver 5.1 Is Receiving Triggers from PsychoPy 3?

Select the "analog" trigger format and trigger line 1-16 are displayed as colored dots/bars. Select the "decimal" trigger format and the triggers are displayed as two groups of 8 bit numbers.
by Coen
Fri Oct 20, 2023 12:52 pm
Forum: BioSemi all other topics
Topic: ActiView TCP server
Replies: 11
Views: 26423

Re: ActiView TCP server

Filtering the TCP outputs is not a standard option in ActiView, it is only possible by modifying the (open source LabVIEW) code. You could consider programming a simple low pass in you C++ client. High frequency interference on the photo sensor signal is usually caused by 100 Hz fluctuating light fr...