I have created a converter to import BioSemi data into ERPLab. I would like to test the converter to ensure each channel is importing the data as it should. Is there a way to create artificial data on the BioSemi intuitive to testing my software? i.e. - is there a way to send a pulse to each individual channel in the array to ensure that the import process is working correctly?
Cheers,
John
Sending pulse to each channel
-
johnwastaken
- Posts: 2
- Joined: Mon May 30, 2011 7:12 pm
- Location: Burnaby, BC
-
johnwastaken
- Posts: 2
- Joined: Mon May 30, 2011 7:12 pm
- Location: Burnaby, BC
Thanks Coen,
There is a further reason why a specific cal-pulse needs to be sent. When importing the data, the software requires you set a gain and a resolution. I need a way to calibrate so I know that the actual amplitude being measured is the same as the amplitude of the data I am analysing. Any ideas?
There is a further reason why a specific cal-pulse needs to be sent. When importing the data, the software requires you set a gain and a resolution. I need a way to calibrate so I know that the actual amplitude being measured is the same as the amplitude of the data I am analysing. Any ideas?
-
Coen
- Site Admin
- Posts: 1164
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
The software should calculate the scaling from the values give in the BDF header, see http://www.biosemi.com/faq/file_format.htm. This leads to LSB = 31.25 nV for the EEG channels. Check your software by making a BDF file with a two bucket test and a calibrated signal generator.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)