TCP/IP processing
MathewS
TCP/IP processing
Post by MathewS »
Hi I am trying to write a small prog to read data from a socket transmitted using your actiview server. I understand in what order the channels are transmitted and that the samples are 3 bytes long and little endian. So:
For 8388607
byte_buffer[0]=-1;//Last element
byte_buffer[1]=-1;
byte_buffer[2]=127;//first element
For -8388608
byte_buffer[0]=0; //Last element
byte_buffer[1]=0;
byte_buffer[2]=-128;//first element
The above is encoded in the byte format so -128=10000000
and 127=11111111
Now how do I translate say 8388607 to a micro volt range or a trigger value or as a value redorded directly to the bdf files?
Mathew
MathewS
Post by MathewS »
Ok after searching the whole forum and analysing the header file I think it is as follows
The value translated is say 8388607
For electrodes
Cal=0.0312
Offset=-0.4844
Foir Status and event signals
Cal=1
Offset=0
On electrodes
So (8388607*Cal)+Offset=261724.054 microvolts ?
The LSB of the 24-bit word is scaled at 1/32th uV. So, a value of 8388607 (2^23-1) is equal to 262144 uV.
Note: on systems for ECG applications, the LSB is scaled at 1/8th uV (check the BDF header)