Page 1 of 1

EPrime Extra Trigger

Posted: Mon Jan 05, 2026 8:12 pm
by maxlab
Hello,

I am having an issue with extra EPrime triggers showing up in BioSemi. Once or twice every 30 trials, two numbers will stack on top of each other as shown in the pictures below. One of the numbers is a repeat of a trigger that happens a second or so later. This seems to have happened spontaneously. I have tried restarting the computer, regenerating the file, disconnecting and reconnecting the BioSemi hardware, checking that RespSwitch = "0" in the configuration file, and changing the triggers to be 1, 2, 4, 8, etc at the advice given to another topic, but none of these have fixed the problem. Is there anything you can think of that may have caused this? What can I do to troubleshoot further?

Thank you for your help!
Max

Re: EPrime Extra Trigger

Posted: Thu Jan 08, 2026 3:41 pm
by Coen
Try sending a zero after each trigger.

Best regards, Coen (BioSemi)

Re: EPrime Extra Trigger

Posted: Fri Jan 09, 2026 8:52 pm
by maxlab
This worked! Thank you!