Simultaneous EEG + EMG

Post Reply
Nathan5653
Posts: 7
Joined: Sat Aug 19, 2023 4:20 pm

Simultaneous EEG + EMG

Post by Nathan5653 »

Hi all,

I am hoping to run an experiment that records EEG and EMG from the hand/arm muscle through the EX channels. For the hand muscle, we would use two of the EX channels but if we were to use a separate CMS/DRL set with flat-type electrodes connected to the AD-box front-panel, would that conflict with the CMS/DRL from the cap?

Thanks,
Nathan
Coen
Site Admin
Posts: 1153
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Simultaneous EEG + EMG

Post by Coen »

You should only connect one CMS/DRL to the AD-box (otherwise, the blue light keeps flashing). Connect either the A set with CMS/DRL, or the front-panel CMS/DRL. So, you can use the EX electrodes with the front-panel CMS/DRL provided the A connector remains unplugged.

Best regards, Coen (BioSemi)
Post Reply