Hi all,
I am hoping to run an experiment that records EEG and EMG from the hand/arm muscle through the EX channels. For the hand muscle, we would use two of the EX channels but if we were to use a separate CMS/DRL set with flat-type electrodes connected to the AD-box front-panel, would that conflict with the CMS/DRL from the cap?
Thanks,
Nathan
Simultaneous EEG + EMG
-
- Site Admin
- Posts: 1153
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Simultaneous EEG + EMG
You should only connect one CMS/DRL to the AD-box (otherwise, the blue light keeps flashing). Connect either the A set with CMS/DRL, or the front-panel CMS/DRL. So, you can use the EX electrodes with the front-panel CMS/DRL provided the A connector remains unplugged.
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)