I am writing a data acquisition client in C/C++ and I am a bit confused on this point. Here: https://www.biosemi.com/faq/make_own_ac ... ftware.htm (at the table on the bottom) the aux channel count appears to be 16 (after EEG and Ex), but here: https://www.biosemi.com/faq/adjust_samplerate.htm it appears to be 7. Then I read posts like this one: https://www.biosemi.nl/forum/viewtopic. ... nsor#p2450 and I see that the GSR channel is read from 2 aux channels (which makes sense as it is a comparative signal) but how do I know which channels to look at? Further confusion comes from the fact that there are 3 AUX ports on the front of the AD box (each having 7 pins). That doesn't square with the 16 channels indicated on the 'How can I make my own acquisition software?' FAQ or the 7 on the 'Is the sampling-rate adjustable on a BioSemi system?' FAQ. So what is correct? Why the discrepancy?
To make a long question short: if I have a sensor plugged into Aux 1 of my AD box, what channels do I need to look at in order to take a reading from the sensor?
Thanks in advance.
Correct way to index AUX channels?
-
- Site Admin
- Posts: 1152
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Correct way to index AUX channels?
The senso signals are sent from the AD-box in a block of 16 channels. Installed sensors generate data on dedicated channels in the block, uninstalled sensors generate flatliners. ActiView processes the 16 channels into 7 sensor channels (displayed and saved to BDF file). Inspect the LabVIEW source code for details about the channel order and processing.
Best regards, Coen (Biosemi)