ecg

Post Reply
rohika
Posts: 4
Joined: Wed Jan 30, 2019 5:27 pm

ecg

Post by rohika »

I usually use BIOSEMI system, but in my next experiment, I want to use only ECG.
However, when I use EXG without any brain channel, ECG doesn't work.
If I use CMS and DRL, ECG works well, but participants need a cap.
Could you please let me know whether ECG without CMS/DRL is available or not?
If yes, please let me know how to do it.

Thanks in advance.
Coen
Site Admin
Posts: 1165
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: ecg

Post by Coen »

CMS/DRL is required for any electrode measurement. When not using the headcap, use a seperate (flat) CMS/DRL connected to the front-panel of the AD-box.

Best regards, Coen (BioSemi)
rohika
Posts: 4
Joined: Wed Jan 30, 2019 5:27 pm

Re: ecg

Post by rohika »

Hi Coen,

Thanks.
Could you please let me know which special cable I should buy?

Sincerely yours,
rohika
Coen
Site Admin
Posts: 1165
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: ecg

Post by Coen »

Order "CMS/DRL Flat Set" (EUR 100 excl. shipping). First check that your AD-box has a CMS/DRL connection on the front-panel (there should be a CMS/DRL label above the most left DIN connector). If not, order "CMS/DRL Flat Set SCSI".

Best regards, Coen (BioSemi)
rohika
Posts: 4
Joined: Wed Jan 30, 2019 5:27 pm

Re: ecg

Post by rohika »

Thanks.
Our AD-box has a CMS/DRL connection.

Is this "CMS/DRL Flat Set" correct? (order code: CMSDRL FLAT)
https://www.biosemi.com/flat_electrode.htm

Sincerely yours,
Rohika
Robbie303
Posts: 28
Joined: Mon Nov 03, 2003 10:03 pm
Location: Amsterdam

Re: ecg

Post by Robbie303 »

CMSDRL FLAT is the ordercode.
There is no image of the CMSDRL FLAT on the website.

For EEG we advise Signa gel.
For all ECG/EMG/EOG measurements we advise Elefix.

The skin on the chest, arms and legs is more difficult to penetrate and make contact with.
That is why for ECG/EMG/EOG, a stronger gel is needed: Elefix.

So, when using TP FLAT electrodes or CMSDRL FLAT electrodes, you need Elefix.
rohika
Posts: 4
Joined: Wed Jan 30, 2019 5:27 pm

Re: ecg

Post by rohika »

Thanks, I will try it.
Post Reply