I usually use BIOSEMI system, but in my next experiment, I want to use only ECG.
However, when I use EXG without any brain channel, ECG doesn't work.
If I use CMS and DRL, ECG works well, but participants need a cap.
Could you please let me know whether ECG without CMS/DRL is available or not?
If yes, please let me know how to do it.
Thanks in advance.
ecg
-
Coen
- Site Admin
- Posts: 1167
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: ecg
Order "CMS/DRL Flat Set" (EUR 100 excl. shipping). First check that your AD-box has a CMS/DRL connection on the front-panel (there should be a CMS/DRL label above the most left DIN connector). If not, order "CMS/DRL Flat Set SCSI".
Best regards, Coen (BioSemi)
Best regards, Coen (BioSemi)
Re: ecg
Thanks.
Our AD-box has a CMS/DRL connection.
Is this "CMS/DRL Flat Set" correct? (order code: CMSDRL FLAT)
https://www.biosemi.com/flat_electrode.htm
Sincerely yours,
Rohika
Our AD-box has a CMS/DRL connection.
Is this "CMS/DRL Flat Set" correct? (order code: CMSDRL FLAT)
https://www.biosemi.com/flat_electrode.htm
Sincerely yours,
Rohika
Re: ecg
CMSDRL FLAT is the ordercode.
There is no image of the CMSDRL FLAT on the website.
For EEG we advise Signa gel.
For all ECG/EMG/EOG measurements we advise Elefix.
The skin on the chest, arms and legs is more difficult to penetrate and make contact with.
That is why for ECG/EMG/EOG, a stronger gel is needed: Elefix.
So, when using TP FLAT electrodes or CMSDRL FLAT electrodes, you need Elefix.
There is no image of the CMSDRL FLAT on the website.
For EEG we advise Signa gel.
For all ECG/EMG/EOG measurements we advise Elefix.
The skin on the chest, arms and legs is more difficult to penetrate and make contact with.
That is why for ECG/EMG/EOG, a stronger gel is needed: Elefix.
So, when using TP FLAT electrodes or CMSDRL FLAT electrodes, you need Elefix.