Only receiving odd (not even) triggers

Post Reply
joshyouarebaker
Posts: 1
Joined: Mon Nov 11, 2013 11:16 am

Only receiving odd (not even) triggers

Post by joshyouarebaker »

I have created a program to send triggers 0 to 255. ActiView fails to detect '1', and then counts only odd numbers (it actually replaces even numbers with the previous odd number. Therefore, when sending triggers 0,1,2,3,4,5,6,7,8,9,10, ActiView displays 3,3,5,5,7,7,9,9.

any advice!?

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Only receiving odd (not even) triggers

Post by Coen »

Line 1 always high (internal pullup), check cable, use analog trigger display (colored bars) for faultfinding.