Hello
I would like to ask what is the recommended procedure and clearing and disinfecting agent for: (1) EEG cap; (2) Cap electrode; (3) EMG flat electrodes; and (4) GSR sensor?
Is it possible to use alcohol?
Thanks
Bori
How to disinfect Biosemi hardware?
Re: How to disinfect Biosemi hardware?
Disinfection is not necessary because it is a non-invasive system (no skin scrubbing is applied to lower skin impedances).
If you insist on disinfection then we recommend:
- alcohol (ethanol or isopropanol) for the electrodes (active electrodes and GSR electrodes)
- mild disinfecting soap for the caps
Thank you very much. Is the percent of the alcohol meters?
Best
Boris
Best
Boris