Hello everybody.
I would like to make an ABR (Auditory Brainstem Response) recording with the use of the Flat Active Electrodes (EXG1-8) in Speedmode 7. What electrode configuration do you recommend (especially where is best place to stick the CMS/DRL pair)?
For the EXG electordes Mastoids, Forhead,Nape and maybe Cz is a good idea?
Hope someone has some experience with ABR & ActiveTwo.
Best regards,
Philipp
CMS+DRL Placement in ABR Recordings
-
- Site Admin
- Posts: 1152
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
We have seen good result with CMS and DRL on the forehead (see http://www.biosemi.com/abr.htm). The loction of CMS and DRL is not critical, see viewtopic.php?t=40 and viewtopic.php?t=33.
Best regards, Coen (BioSemi)