When I hook the ActiView batterybox to the amplifier the POWER / CM in range / low battery blink 3 times very quickly. The amplifier's POWER button glows for a quick moment, then the SHUTDOWN light stays on.
Then the software throws this error:
Error 5000 occured at No Incoming Data
Possible reason(s):
ADC-box power is off
or
Optical fiber is disconnected
Please advise,
battery box shutdown
-
- Site Admin
- Posts: 1152
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Is the (red) shutdown light on the batteries ON after shutdown?
If yes, then please sent the batteries for service. If no, (green power light stays ON, but AD-box shuts down), then you probably have a problem with electrode connectors polluted with water and/or gel. Clean the connectors on the electrode sets with distilled water and dry with alcohol. If that does not solve the problem, then please sent the AD-box and electrode sets for service. Do not attempt to clean the electrode connectors on the AD-box yourself
Best regard, Coen (BioSemi)
If yes, then please sent the batteries for service. If no, (green power light stays ON, but AD-box shuts down), then you probably have a problem with electrode connectors polluted with water and/or gel. Clean the connectors on the electrode sets with distilled water and dry with alcohol. If that does not solve the problem, then please sent the AD-box and electrode sets for service. Do not attempt to clean the electrode connectors on the AD-box yourself