I compared the data obtained via Matlab/EEGLAB from the BDF files and the data coming from the "Downsample.vi" for the EXi channels and found a difference by factor 10000 (probably 8196 ?).

What is the exact reason for that and where does the scaling happen in the full Actiview version ? I need comparable data for the investigations I have done offline and which I will do online using modified Actiview.

I ve found a scaling by 8196 caused by the variable ElecGain, but that appears to me as if it only affects the displaying. Is there sth. similar affecting the savings as well ?
Back to top
View user's profile Send private message
Coen
Site Admin


Joined: 26 Mar 2004
Posts: 399
Location: Amsterdam, Netherlands

PostPosted: Tue Jun 22, 2010 3:49 pm    Post subject: Reply with quote

The 32-bit data words in the output array from Downsample.vi have a scaling of LSB = 1/8192 uV (1/32 * 1/256)

The 24-bit data words in the BDF file have an scaling of LSB = 1/32 uV

The Elecgain variable changes both the above LSB values for a few special versions of the ActiveTwo with passive electrodes. With standard systems with active elctrodes, Elecgain = 0 (set by the CFG file), and the above LSB values apply.
The above LSB values are not affected by the selected display gain in ActiView.
The above LSB values are valid for the active electrode channels, some of the sensor channels have different scaling (see BDF header).

Best regards, Coen (BioSemi)
Back to top
View user's profile Send private message Send e-mail Visit poster's website
Display posts from previous:   
Post new topic   Reply to topic    BioSemi.nl Forum Index -> BioSemi Software (LabVIEW programming) All times are GMT + 1 Hour
Page 1 of 1

 
Jump to:  
You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum
You cannot vote in polls in this forum


Powered by phpBB © 2001, 2005 phpBB Group
Protected by Anti-Spam ACP