ABR / ERP simultaneous recording
Posted: Thu May 12, 2011 8:57 pm
by Emz
We are attempting to collect both ABRs and ERPs simultaneously, by recording raw data then offline filtering and post-processing.
We started using the ABR mode and got fairly good results, although our ERPs appeared attenuated and we suppose this was due to the 100Hz high pass filter.
Some colleagues helped us to set up a standard mode recording using the external ports for our 1 data electrode and 2 reference electrodes. We used speedmode7 (though to be honest I'm not entirely sure what the implications of that are). We are still fiddling with our data, but it looks rather noisy.
Is it possible to collect good data for both ABRs and ERPs simultaneously, and if so, what parameters/set up would you recommend?
Emz
Posted: Fri May 13, 2011 2:32 pm
by Coen
We did not foresee that customers would want to perform simultaneous
ABR and ERP. Therefore, the current firmware and acquisition software does not support simultaneous
ABR and ERP.
The possibilities with your current system are:
- Speedmode 9: record
ABR only from the special
ABR electrodes with the special
ABR acquisition software (ActiABR-101)
- Speedmodes 4-7: record with standard active electrodes according to
http://www.biosemi.com/faq/adjust_samplerate.htm with standard ActiView.
In speedmode 7, you can acquire channels 1-32 and the 8 EX channels and the sensor channels on 16 kHz. In speedmode 7, the special
ABR channels are disabled. It is possible to use 2 standard EX electrodes to record
ABR in this speedmode, see
http://www.biosemi.com/abr.htm. However, with stadnard EX electrodes the (high frequency) signal-to-noise ratio may be worse than with the special
ABR electrodes (because the standard EX electrodes and channels are optimized for lower EEG frequencies, whereas the
ABR electrodes and channels are optimized for high frequencies).
Recording the special
ABR electrodes together with a number of standard electrodes is technically possible, but would require custom made firmware and acquisition software. Please contact BioSemi directly to discuss the requirements and possibilities.
Best regards, Coen (BioSemi)
Recording ABRs, cABRs and MLRs in speed mode 7
Posted: Mon Jun 13, 2011 4:41 pm
by tom_campbell75
Hi,
I hope that this article is of interest:
http://sccn.ucsd.edu/eeglab/refs/Campbell11.pdf
Best regards,
Tom.
Thank you
Posted: Fri Jul 08, 2011 7:12 pm
by Alexandre
Thank you for this fresh paper. So if I understand well you actually used mode 7 with conventional Biosemi electrodes ? Did you carry out your analyses with EEGLAB ? What about introducing a random jitter between two stimuli to avoid responses from previous trials ?
Best
Alexandre
Posted: Fri Jul 15, 2011 11:49 pm
by tom_campbell75
Hi Alexandre, Yes Mode 7 with the conventional pin electrodes. Sure, I analysed it in MATLAB. Jittering and Adjarring is in the pipeline. T.
Re: ABR / ERP simultaneous recording
Posted: Thu Oct 12, 2017 1:35 pm
by sarahah