Conductance gel

Posted: Tue Jul 17, 2007 1:18 am
by partylab
Hi

I'm looking into trying some different types of conductance gel. Currently our lab uses ECI Electro-Gel (thats the electro cap product).

I'm curious to know if other conductance gels may have higher saline levels that would make it easier to bring impedances down? Our current work with children requires us to have impedances below 5 kohms. Usually this requires manipulating the gel/electrode with a stick....which we would like to stop doing.

Any suggestions?

pl

Posted: Tue Jul 17, 2007 11:01 pm
by Coen
With BioSemi active electrodes, it is not necessary to reduce the impedances by scrubbing. The signal quality is optimal without any skin preparation.

Best regards, Coen (BioSemi)

Posted: Tue Mar 03, 2009 9:16 pm
by smcphers
We use Signa Gel which has a 0.5% saline base which works quite well.

Posted: Tue Mar 03, 2009 10:09 pm
by terracop
There are several gels which will do the job. Signa is fine. But most interesting is to know how do you measure impedances and why?

A way to monitor correct wiring

Posted: Wed Jul 08, 2009 5:03 pm
by Alexandre
[quote]most interesting is to know how do you measure impedances and why?[/quote]

Indeed I understood that because of active electrodes, no scrubbing is necessary, but still, how can i control that the contact between the electrodes and the skin is good enough (especially in the case of using paste for electrodes outside cap (i.e. ARB recording) ?

Thanks in advance for your reply

Posted: Thu Jul 09, 2009 11:52 am
by Coen
The electrode offset gives a good indication of the contact quality. The offset should be stable, and within the +50 mV and -50 mV range.

The ABR software offers an additinal (and best possible) check of the electrode contact quality: you can actually measure the electrode noise level with the online averager (gray traces in the graphs).

Best regards, Coen (BioSemi)

Posted: Thu Jul 09, 2009 4:13 pm
by Alexandre
Thanks, I will try this out !
Best