Hi
I'm looking into trying some different types of conductance gel. Currently our lab uses ECI Electro-Gel (thats the electro cap product).
I'm curious to know if other conductance gels may have higher saline levels that would make it easier to bring impedances down? Our current work with children requires us to have impedances below 5 kohms. Usually this requires manipulating the gel/electrode with a stick....which we would like to stop doing.
Any suggestions?
pl
Conductance gel
A way to monitor correct wiring
[quote]most interesting is to know how do you measure impedances and why?[/quote]
Indeed I understood that because of active electrodes, no scrubbing is necessary, but still, how can i control that the contact between the electrodes and the skin is good enough (especially in the case of using paste for electrodes outside cap (i.e. ARB recording) ?
Thanks in advance for your reply
Indeed I understood that because of active electrodes, no scrubbing is necessary, but still, how can i control that the contact between the electrodes and the skin is good enough (especially in the case of using paste for electrodes outside cap (i.e. ARB recording) ?
Thanks in advance for your reply
-
- Site Admin
- Posts: 1152
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
The electrode offset gives a good indication of the contact quality. The offset should be stable, and within the +50 mV and -50 mV range.
The ABR software offers an additinal (and best possible) check of the electrode contact quality: you can actually measure the electrode noise level with the online averager (gray traces in the graphs).
Best regards, Coen (BioSemi)
The ABR software offers an additinal (and best possible) check of the electrode contact quality: you can actually measure the electrode noise level with the online averager (gray traces in the graphs).
Best regards, Coen (BioSemi)