ECG/EEG/EMG is measured BETWEEN two (or more) electrodes. The signal of a single electrodes with respect to CMS (raw, unreferenced) is not a valid signal, see
https://www.biosemi.com/faq/cms&drl.htm
viewtopic.php?f=1&t=33&p=86
viewtopic.php?f=1&t=40&p=109
viewtopic.php?f=1&t=177
So, for ECG use (at least) two electrodes on the chest and subtratc the signals in sofwtare.
The scaling for active electrode systems is:
32-bit words from USB-reciever: divide by 8192 (32*256) to convert to microVolt
24-bit words in the BDF file: divide by 32 to convert to microVolt
Note: custom systems for passive electrodes have 1/4th of the above gain (divide by respectively 32768 and 128)
Note that the system uses a DC amplifier. Therefore, the ECG/EEG/EMG signal is superimposed on a relatively large (typically several tens of milliVolt) electrode offset voltage. Use a high-pass filter or baseline correction to cancel the DC offset and extract the biopotentials.
Best regards, Coen (BioSemi)