Dear,

I was wondering if it is possible to extract somehow the status information about the electrode offset and the CMS in range from the BDF files? I know it is saved somewhere (as I can see it an the BDF reader), but I don't know how to extract this information into a readable file format (for R or Matlab). Does anyone have an idea for that?

Thank you,
Katharina

Coen
Site Admin
Posts: 899
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Extracting Status Information (Electrode Offset, CMS in range)

Post by Coen » Mon Nov 20, 2017 4:49 pm

Status information is coded in the Most Significant Byte (8 Most Significant Bits) of the trigger/status channel (always the last channel), see https://www.biosemi.com/faq/trigger_signals.htm

Offsets are the DC values of the channels, see viewtopic.php?t=56

Best regards, Coen (BioSemi)