Drifting of record during experiment

Post by Johan_P »

Hi all,
I'm using biosemi active2 system with 16 electrodes + 8 external.
I'm doing CNV experiment and manage to obtain quite good result but I lose a lot of data due to the drifting of the active electrode. It's not a very big drifting but my baseline is not on zero and not stable. So for some stimulation, I have drifting through the positive values and the CNV activity (which should be in the negative value) is lost in this drifting. I can't take these kind of data into account.

Could you have some tips regarding this problem ? Do you know what could be the origin? I know that temperature, perspiration, stress could change a bit the drifting, but there also could be other more controllable factor that I don"t know...

Thanks,
Johan

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Drifting of record during experiment

Post by Coen »

Raw data on file is always DC (regardless of display filters during recording), apply an high-pass filter during analysis.

Best regards, Coen (BioSemi)

Johan_P
Posts: 2
Joined: Tue Sep 23, 2014 4:07 pm

Re: Drifting of record during experiment

Post by Johan_P »

Thanks Coen,
I already use a passband filter between 0.03Hz to 30Hz. The drifting is still here.
For example, after filtering, cleaning and epoching you can obtain something like the picture below


What do you think ?

Johan
Attachments
E10S02_EB14B.jpg
E10S02_EB14B.jpg (14.29 KiB) Viewed 15294 times

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Drifting of record during experiment

Post by Coen »

Excessive drift is usually caused by perspiration. Try organize a cool and comfortable environment for the subjects during the experiment.

Best regards, Coen (BioSemi)