Only receiving odd (not even) triggers

Post by joshyouarebaker » Mon Jun 02, 2014 3:39 pm

I have created a program to send triggers 0 to 255. ActiView fails to detect '1', and then counts only odd numbers (it actually replaces even numbers with the previous odd number. Therefore, when sending triggers 0,1,2,3,4,5,6,7,8,9,10, ActiView displays 3,3,5,5,7,7,9,9.

any advice!?

Coen
Site Admin
Posts: 924
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Only receiving odd (not even) triggers

Post by Coen » Mon Jun 02, 2014 6:32 pm

Line 1 always high (internal pullup), check cable, use analog trigger display (colored bars) for faultfinding.