Event related average problem

Post Reply
smar
Posts: 1
Joined: Thu Dec 15, 2011 12:44 pm
Location: Switzerland

Event related average problem

Post by smar » Fri Jan 06, 2012 5:56 pm

Hi,
we're trying to use the Event related average module but without much success: even if the right speedmode is set, channels and trigger are correctly selected, no traces appear when pressing the 'Start Avg'. Could you please help me with this issue?
Thanks a lot in advance!

Coen
Site Admin
Posts: 969
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Post by Coen » Mon Jan 09, 2012 2:15 pm

Is the acquisition running and do the triggers appear in the monopolar display (bottom of page) ?

Is the REJECTED SWEEPS counter counting ? If yes, then relax the conditions of the on-line artifact rejection. If no, then check the triggers settings (triggers are apparently not detected by the avg modules.

Best regards, Coen (BioSemi)

Post Reply