Irregular noise all over
Irregular noise all over
Hi, what could be the source of this type of noise?
- Attachments
-
- noise.jpg (248.35 KiB) Viewed 23338 times
-
- Site Admin
- Posts: 1152
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Irregular noise all over
P4 does not make contact with the skin (not enough gel), or is defect (test in water).
Best regard. Coen (BioSemi)
Best regard. Coen (BioSemi)
Re: Irregular noise all over
Hi,
Thanks for this answer! I apologise for my totally beginner's questions, but with one of our subjects, the signal was terrible (see the picture) and we had to let the person go home. I checked the offset, the majority of electrodes was around -10 to -30 mV, but, they were all unstable (they were jittering). I of course tried adding more gel a couple of times and checking again. This would sometimes decrease the offset to -10 to -20 mV, but the electrode would still be unstable. By the way, the blue light was on and all the electrodes passed the bucket test.
Since we previously ran subjects in the same room, with the same setting, and didnt have these issues, it makes me think that there might be something specific to this subject (e.g. thick hair) - do you think this is the case? And can we do anything to improve the signal in those situations? Or does this look like something else, am I missing something?
Thanks,
Bojana
Thanks for this answer! I apologise for my totally beginner's questions, but with one of our subjects, the signal was terrible (see the picture) and we had to let the person go home. I checked the offset, the majority of electrodes was around -10 to -30 mV, but, they were all unstable (they were jittering). I of course tried adding more gel a couple of times and checking again. This would sometimes decrease the offset to -10 to -20 mV, but the electrode would still be unstable. By the way, the blue light was on and all the electrodes passed the bucket test.
Since we previously ran subjects in the same room, with the same setting, and didnt have these issues, it makes me think that there might be something specific to this subject (e.g. thick hair) - do you think this is the case? And can we do anything to improve the signal in those situations? Or does this look like something else, am I missing something?
Thanks,
Bojana
- Attachments
-
- yy.jpg (229.56 KiB) Viewed 23233 times
-
- Site Admin
- Posts: 1152
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
Re: Irregular noise all over
You probably added so much gel that there was a gel bridge between CMC and DRL (40 dB less CMRR). Apply just enough gel to fill the holder, but not more. Use the tip of the syringe to work the gel between the hair.
Offsets should be between approx. +100 mV and -100 mV and stable. If within this range, then closer to zero is not better.
Best regards, Coen (BioSemi)
Offsets should be between approx. +100 mV and -100 mV and stable. If within this range, then closer to zero is not better.
Best regards, Coen (BioSemi)
Re: Irregular noise all over
Hi,
Thanks a lot for this answer. Yes, it was probably the gel.
Regarding the stability - is the way to improve it adding more gel (unless it is too much, of course)?
Bojana
Thanks a lot for this answer. Yes, it was probably the gel.
Regarding the stability - is the way to improve it adding more gel (unless it is too much, of course)?
Bojana