Triggering with BioSemi Active 2

Post Reply
wmeinhold
Posts: 1
Joined: Tue Oct 18, 2016 6:19 pm

Triggering with BioSemi Active 2

Post by wmeinhold »

Hi, I am new to this system, and have a question about the trigger inputs.
The trigger box sets the input pins to 3.3V when nothing is connected.
As far as I can tell, the only way to change this is by shorting the inout so that the voltage across the pins goes to 0.
I would like to trigger off of the signal from an accelerometer, or really just any signal, is it possible to do this?

Thank You

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: Triggering with BioSemi Active 2

Post by Coen »

The trigger inputs have an internal 10 kOhm pull up resistor. Your trigger signal must be TTL level (Low = 0V, High is 3.3-5V), and must be able to drive the 10 kOhm resistor (be able to pull the input below approx. 1V).

Best regards, Coen (BioSemi)