AD box won't turn on

Post Reply
cogdeveeg
Posts: 1
Joined: Sun Apr 24, 2016 6:42 pm

AD box won't turn on

Post by cogdeveeg »

After plugging in everything (ground, facial electrodes, and bundles) into the AD box, the AD box wouldn't turn on, even though the battery was completely fine. The AD box would initially have the CM in range light would flash twice then turn off, and the green light would stay on for just as long before turning off. We tried changing out the battery, and unplugging and plugging everything back in, turning it on while nothing was in, and nothing seemed to help. We checked the optic fiber to make sure it was connected as well, but this didn't help either.

Now, the CM range light won't stay on (blinking), but both green lights are continuous.

What's wrong, and how can we fix the issue(s)? This has happened in the past, but doesn't happen every time.

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: AD box won't turn on

Post by Coen »

Shutdown of the AD-box (with battery turned on) is caused by activation of the internal safety circuits. Most probably cause are connectors polluted with electrode gel (causing a short circuit). Consult your local representative for further advise.

Best regards, Coen (BioSemi)

Post Reply