Page 1 of 1

How to disinfect Biosemi hardware?

Posted: Wed Nov 27, 2013 8:29 pm
by Boris
Hello

I would like to ask what is the recommended procedure and clearing and disinfecting agent for: (1) EEG cap; (2) Cap electrode; (3) EMG flat electrodes; and (4) GSR sensor?

Is it possible to use alcohol?

Thanks

Bori

Re: How to disinfect Biosemi hardware?

Posted: Thu Nov 28, 2013 11:42 am
by Coen
Disinfection is not necessary because it is a non-invasive system (no skin scrubbing is applied to lower skin impedances).

If you insist on disinfection then we recommend:

- alcohol (ethanol or isopropanol) for the electrodes (active electrodes and GSR electrodes)
- mild disinfecting soap for the caps

Best regards, Coen (BioSemi)

Re: How to disinfect Biosemi hardware?

Posted: Thu Nov 28, 2013 11:45 am
by Boris
Thank you very much. Is the percent of the alcohol meters?

Best

Boris

Re: How to disinfect Biosemi hardware?

Posted: Thu Nov 28, 2013 12:32 pm
by Coen
No