How to disinfect Biosemi hardware?

Post Reply
Boris
Posts: 12
Joined: Fri Apr 09, 2010 10:29 am
Location: Israel

How to disinfect Biosemi hardware?

Post by Boris »

Hello

I would like to ask what is the recommended procedure and clearing and disinfecting agent for: (1) EEG cap; (2) Cap electrode; (3) EMG flat electrodes; and (4) GSR sensor?

Is it possible to use alcohol?

Thanks

Bori

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: How to disinfect Biosemi hardware?

Post by Coen »

Disinfection is not necessary because it is a non-invasive system (no skin scrubbing is applied to lower skin impedances).

If you insist on disinfection then we recommend:

- alcohol (ethanol or isopropanol) for the electrodes (active electrodes and GSR electrodes)
- mild disinfecting soap for the caps

Best regards, Coen (BioSemi)

Boris
Posts: 12
Joined: Fri Apr 09, 2010 10:29 am
Location: Israel

Re: How to disinfect Biosemi hardware?

Post by Boris »

Thank you very much. Is the percent of the alcohol meters?

Best

Boris

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: How to disinfect Biosemi hardware?

Post by Coen »

No

Post Reply