Battery Issue

Posted: Wed Oct 02, 2013 2:38 pm
by vanv0076
Hello,

We've recently been having issues in our lab with a battery that does not appear to be working correctly. Whenever the battery is turned on the green light indicating that it is powered on does not activate, nor does the collection box indicate that it is receiving power from the battery when it is connected to the battery and the battery is turned on. When the battery is connected to a charger the green light on the charger labeled "Ready" stays lit, it doesn't give the yellow "Charging" light. This has been tested on two different chargers, both of which have no issues charging other batteries. Is there anything we can do about this issue?

Thanks,
Abe

Re: Battery Issue

Posted: Wed Oct 02, 2013 4:56 pm
by Coen
Check the fuse. If the fuse is OK, then return it to replace the internal cells.

Best regards, Coen (BioSemi)

Re: Battery Issue

Posted: Wed Nov 13, 2013 3:21 pm
by vanv0076
Replacing the fuse seems to have solved the problem. Thanks again!

Re: Battery Issue

Posted: Mon Nov 18, 2013 9:07 pm
by vanv0076
It looks like I was mistaken. Replacing the fuse did change the situation, now the green light does activate when the battery is turned on but the light on the collection box does not illuminate to indicate that it is receiving power when it is connected to the battery. The collection box works just fine with the other battery that we have. Are there any possible fixes for the situation that we could do on our own or should we send it in?

Thanks again,
Abe

Re: Battery Issue

Posted: Tue Nov 19, 2013 12:18 pm
by Coen
Open the blue power connector and check for loose connections (most probable cause of your issue). If all 3 wires are properly connected, then sending the battery to us for service is the next step.

Best regards, Coen (BioSemi)

Re: Battery Issue

Posted: Wed Dec 11, 2013 3:47 pm
by vanv0076
Thanks for the tip. We fixed it after noticing one of the wires was not connected.