mgrivich
Posts: 3
Joined: Sat Aug 24, 2013 12:22 am

256 electrodes + analog input box

Post by mgrivich »

Greetings,

I configured our setup as described by http://www.biosemi.com/aib.htm. There should be 256 active EEG electrodes as well as 32 active analog input electrodes. It does not work as advertised. Below I describe what happens, as well as two other configurations (to verify that the boxes have basic functionality) . I am inspecting the data using ActiView 7.05. Please advise. Thanks.

AD (Speed Mode 4) - AIB (Speed Mode 8) - Receiver
2048 Hz
Channel 1-32: AIB
Channels 33-256: zero
Missing all AD channels

AD (Speed Mode 4) - Reciever
2048 Hz
Channel 1-264: AD
AIB is not connected

AIB (Speed mode 0) - AD (Speed mode 1) - Receiver
2048
AD-Box 1: Channels 1-32: AIB
Channels 33-128+external: zero
AD-Box 2: Channels 1-128+external: AD
AD channels 129-256 are missing.

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: 256 electrodes + analog input box

Post by Coen »

AD (Mode 4) --> AIB (Mode 8) --> Receiver, which speedmode is indicated in ActiView ("about" tabpage) ?

AIB (Mode 4) --> Receiver, which speedmode is indicated in ActiView ?

Best regards, Coen (BioSemi)

mgrivich
Posts: 3
Joined: Sat Aug 24, 2013 12:22 am

Re: 256 electrodes + analog input box

Post by mgrivich »

>AD (Mode 4) --> AIB (Mode 8) --> Receiver, which speedmode is indicated in ActiView ("about" tabpage) ?

AD-box Mk1, Speed Mode 0. The physical switches are in the proper position. The software is not reporting it correctly.

>AIB (Mode 4) --> Receiver, which speedmode is indicated in ActiView ?

AD-box Mk2, Speed Mode 4

Coen
Site Admin
Posts: 1124
Joined: Fri Mar 26, 2004 7:00 pm
Location: Amsterdam, Netherlands
Contact:

Re: 256 electrodes + analog input box

Post by Coen »

Looks like a defect AIB (probably a bad speedmode rotary switch), please send the AIB to your local BioSemi representative for service.

Best regards, Coen (BioSemi)

mgrivich
Posts: 3
Joined: Sat Aug 24, 2013 12:22 am

Re: 256 electrodes + analog input box

Post by mgrivich »

Will do. Thanks.