Hello!
This is a very basic question...
We are learning to use EEG 128 electrodes, and there is something which is unclear...Could anyone help us?
We've been told that Electrodes A +CMS/DRL have to be connected to the box before the other group of Electrodes (B, C and D). Is it right? If we don't proceed like that, will EEG recording be "wrong" ? And why?
Waiting for your aswers...
laura
CMS/DRL:basic question
-
- Site Admin
- Posts: 1125
- Joined: Fri Mar 26, 2004 7:00 pm
- Location: Amsterdam, Netherlands
- Contact:
It does not matter in which order the electrodes connectors are plugged into the AD-box. However, valid EEG signals can only be measured with the CMS and DRL electrodes properly connected to the subject (blue light continuously on, see viewtopic.php?t=42)
Best regards, Coen (BioSemi)