Massive 16Hz contamination from GSR sensor

Posted: Mon May 28, 2012 8:10 pm
by Alexandre
Hello,

We are willing to record GSR alongside EEG, we are using the 16Hz SC coupler.
When connected to the subjects finger, I can see a good GSR signal and how it is modified by subjects taking a deep breath.
But the EEG channels are all massively contaminated by 16Hz (barbwire look to quote from another topic in this forum), so I had to disconnect it for my experiments. I tried to follow the guidelines I read but with no luck.

Any suggestion ?

Thanks,

Alex

Re: Massive 16Hz contamination from GSR sensor

Posted: Tue May 29, 2012 10:28 am
by Coen
If ALL channels are equally contaminated, then you have a Common Mode problem. Probable causes are referencing errors, and/of suboptimal mounting of the CMS/DRL electrodes. Please contact your local BioSemi representative for specific help.

Best regards, Coen (BioSemi)